This Information Applies To: Agilent OpenLab CDS 2.x controlling Agilent LC Systems
Issue
This article outlines an instruction on how to submit a shutdown method on OpenLab CDS 2.3 or above.
Background
Submit a shutdown run to the run queue that will shut down Agilent LC instrument after all other runs in the run queue are complete.
Steps to Follow
- Create an acquisition method (.amx file) to use it as a shutdown method. (See Figure 1)
See Table 1 for suggested instrument parameters for the LC instrument shutdown method.
Figure 1: Create a shutdown amx file
Note: The following recommendations are just a general guideline. Method conditions may be different according to the application, installed column, and instrument configuration.
Table 1: Instrument Parameters for LC shutdownPump Flow: 0 mL/min, or a low flow compatible with your application
Stoptime: 1 min
Column Oven Temperature left: Ambient
Temperature right: combined
Stoptime: As Pump/ Injector
DAD/ MWD/ VWD In the Advanced section, go to the Lamps on required for acquisition, and clear the UV lamp check box
Stoptime: As Pump/ Injector
FLD Signal: clear the Acquire Signal check box
In the Special setpoints section, go to Lamp and clear Lamp on required for analysis check box
Stoptime: As Pump/ Injector
RID Optical Unit Temperature: Not controlled, or the same temperature as the analytical method to allow rapid system equilibration for next run
Signal: Clear the Acquire Signal check box
Stoptime: As Pump/ Injector
ELSD Stoptime: As Pump/ Injector
Switch off: deselect Switch off at the end of the run; deselect Switch off on program exit
For G4260B, G4261B, G7102A: Output Signal: deselect Acquire Signal
Tips: If buffer is used in the mobile phase, flush the system with a solvent containing water instead of buffer, according with your method and column compatibility. - Click Browse to select the shutdown acquisition method (Figure 2).
The shutdown method can only be submitted in the Run Queue. - Select the acquisition method and click Open. (See Figure 3)
If the method is obsoleted, then the shutdown run cannot be submitted.
- Click Submit Shutdown Run. (See Figure 4)
- The shutdown run is added to the end of the run queue and is highlighted in yellow. You cannot submit another shutdown run until the submitted shutdown run is complete. (See Figure 5)
Figure 5: Shutdown Method in Pending
Confirmation: Once the shutdown run begins, the shutdown process is recorded in the Activity Log. (See Figure 6)
|