Chromatography Software Articles How to Submit a Shutdown Method in OpenLab 2.x
  • Chromatography Software Articles
  • Chromatography Software Videos
  • Chromatography Software Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Chromatography Data Systems
    • +OpenLab CDS Administration
    • -OpenLab CDS Acquisition
      • How to Create, Save, and Open Acquisition Methods in Agilent OpenLab CDS Software
      • How to Submit a Single Sample in Agilent OpenLab CDS (ver 2.x)
      • How to Create, Edit and Run a Sequence in Agilent OpenLab CDS
      • How to Submit a Shutdown Method in OpenLab 2.x
      • How to Create and Submit an LC Sample Preparation Method in OpenLab CDS 2.x
      • How to Check the Audit Trail of an Acquisition Method in OpenLab CDS (ver. 2.x) System
      • Print modules information: serial number, firmware revision, driver version, and model part number
      • GC Column Configuration Using OpenLab CDS Software
    • +OpenLab CDS Data Analysis
    • +OpenLab CDS Troubleshooting

How to Submit a Shutdown Method in OpenLab 2.x

Created by Linh Dinh Linh Dinh 14 days ago | Last modified by Carlos Vargas Carlos Vargas 13 days ago

This Information Applies To: Agilent OpenLab CDS 2.x controlling Agilent LC Systems


Issue

This article outlines an instruction on how to submit a shutdown method on OpenLab CDS 2.3 or above.


Background

Submit a shutdown run to the run queue that will shut down Agilent LC instrument after all other runs in the run queue are complete.

Steps to Follow

  1. Create an acquisition method (.amx file) to use it as a shutdown method. (See Figure 1)
    See Table 1 for suggested instrument parameters for the LC instrument shutdown method.




    Figure 1: Create a shutdown amx file


     
    Note: The following recommendations are just a general guideline. Method conditions may be different according to the application, installed column, and instrument 
    configuration.


    Table 1
    : Instrument Parameters for LC shutdown 

    Pump

    Flow: 0 mL/min, or a low flow compatible with your application

    Stoptime: 1 min

    Column Oven

    Temperature left: Ambient

    Temperature right: combined

    Stoptime: As Pump/ Injector

    DAD/ MWD/ VWD

    In the Advanced section, go to the Lamps on required for acquisition, and clear the UV lamp check box

    Stoptime: As Pump/ Injector

    FLD

    Signal: clear the Acquire Signal check box

    In the Special setpoints section, go to Lamp and clear Lamp on required for analysis check box

    Stoptime: As Pump/ Injector

    RID

    Optical Unit Temperature: Not controlled, or the same temperature as the analytical method to allow rapid system equilibration for next run

    Signal: Clear the Acquire Signal check box

    Stoptime: As Pump/ Injector

    ELSD

    Stoptime: As Pump/ Injector

    Switch off: deselect Switch off at the end of the run; deselect Switch off on program exit

    For G4260B, G4261B, G7102A: Output Signal: deselect Acquire Signal


     Tip 
    Tips: If buffer is used in the mobile phase, flush the system with a solvent containing water instead of buffer, according with your method and column compatibility.

  2. Click Browse to select the shutdown acquisition method (Figure 2).
    The shutdown method can only be submitted in the Run Queue.


     Figure 2. Browsing a Shutdown Method

  3. Select the acquisition method and click Open. (See Figure 3)
    If the method is obsoleted, then the shutdown run cannot be submitted.


     Figure 3: Open a Shutdown Method

  4. Click Submit Shutdown Run. (See Figure 4)


     Figure 4: Submit Shutdown Run

  5. The shutdown run is added to the end of the run queue and is highlighted in yellow. You cannot submit another shutdown run until the submitted shutdown run is complete. (See Figure 5)


     Figure 5: Shutdown Method in Pending


     Success 
    Confirmation: Once the shutdown run begins, the shutdown process is recorded in the Activity Log. (See Figure 6)


 Figure 6: Activity Log

 Tip 
Learn how to effectively operate your Agilent LC System:
HPLC-OLII-1122e - Agilent OpenLab CDS for HPLC - Instrument Status and Single Runs
HPLC-OLII-1122eV2 - Agilent OpenLab CDS for HPLC - Instrument Status and Single Runs (Ver. 2.6) e-learning courses available from Agilent education

  • OpenLab Chromatography Data System
  • 1220
  • 1100 series
  • OpenLab CDS
  • 1290 infinity
  • 1260 Infinity
  • 1260 Infinity II
  • 1200 Series
  • 1120
  • Analytical LC
  • 1290 Infinity II
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help