Locate the current tune file in Windows File Explorer under the path
x:\Masshunter\GCMS\n\5977
(x is the installation drive of MassHunter folder, n is the instrument number). Rename theatune.u
tune file ( e.g.atune_old.u
). -
Locate and Copy the default tune file (
atune.u
) using Windows File Explorer from pathC:\GCMS\MSSETUP\5977,
and paste it into the same location as the renamed tune file in Step 1. -
The copied default tune file needs the MS polarity set in Manual Tune view (View>Tune and Vacuum Control > Parameters > Manual Tune). Save the tune file with the correct polarity for your instrument. The polarity information will be in the previous tune reports but also confirm it on the instrument.
In Figure 1, see the “X” symbol on the right side of the “Neg” box on the MS analyzer label – it means that negative polarity is to be set for this instrument.
Figure 1. MS Analyzer label – Negative Polarity
-
Tune the MS before data acquisition after resetting the EI tune file (
atune.u
) to the default values.
(atune.u
) to the default values.- Close the GC/MS Acquisition software and open the Agilent GC/MS Configuration editor.
- Click the instrument number on the left at the top to open the instrument configuration.
- Do not change any information, just click Ok, when the dialog box opens "proceed with update to instrument configuration," click Yes.
- The update is now done, close the Agilent GC/MS Configuration editor.
- Tune the MS before data acquisition after resetting the EI tune file (
atune.u
) to the default values.
|
Learn how to effectively operate your Agilent GC/MS MassHunter software:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning course available from Agilent education
|