Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Mass Spectrometry Software Portal
  • Agilent Knowledge Portal
Mass Spectrometry Software Portal
Articles How to Reset the Agilent GC/MSD Tune File to Default Values
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Mass Spectrometry Software Portal Home
  • Mass Spectrometry Software Portal Articles
  • Mass Spectrometry Software Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  25 Jul 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC/MS Systems Mass Spectrometry Software
  • Component :  GC/MS Data Acquisition Software GC/MS Single Quadrupole
  • Product Name :  5973 Series GC/MSD 5975 Series GC/MSD 5977 Series GC/MSD GC/MSD MassHunter Acquisition Software
  • Task :  Data Acquisition Troubleshooting
Related Articles
  • MSD Fault: RFPA Difficulty

    This Information Applies To: Agilent 5975 and 5977 Series GC/MSD systems. Issue During tuning, MassHunter prompted the error message: "MSD Fault- RFPA Difficulty", and tune action stopped. Steps to Follow Try to reconfigure the GC/MS. See article: How To Configure a New GC/MSD with the Agilent GC/MS Configuration Editor . If autotune still failed, try to tune starting with default tune, as a corrupted tune file could be one of the reasons for the failure. See the article: How to...
  • How to Check the Air and Water Background in the Agilent 5975 and 5977 GC/MSD

    This Information Applies To: Agilent 5975 and 5977 GC/MSD systems with MassHunter Acquisition software Issue: High-quality GC/MS data requires a low, and stable background of air and water present in the MS analyzer. This article describes how to check the air and water background in the Agilent 5977 and 5975 GC/MSD. Steps to follow: Introduction: Air and Water background will always be present in the GC/MS vacuum system. See Figure 1 , Figure 2 and Figure 3 . Air and Water background...
  • Reviewing the Autotune Report for Agilent 5977 GC/MSD Systems in EI Mode

    This Information Applies To: All Agilent 5977 GC/MSD systems working in EI mode Issue: You can evaluate the 5977 MSD tune performance by reviewing the autotune report. Follow the steps in this article to review MSD tune performance. Steps to follow: 1. Obtain the hardcopy tune report or open the PDF tune report file in the path X/MassHunter/GCMS/1/5977/PDF (see Figure 1 ). Figure 1. PDF tune file in X/MassHuner/GCMS/1/5977/PDF 2. The mass peak profiles for the tune...
  • How To Configure a New GC/MSD with the Agilent GC/MS Configuration Editor

    Multiple Results
    This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Issue The GC/MS Configuration Editor can be useful when troubleshooting GC/MS MassHunter Acquisition software. It may be useful when troubleshooting some types of software...
    • Found In:
    • Articles
    • ,
    • Articles
  • What to do if my LC-QQQ Autotune Fails?

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems Issue Autotune fails to complete. Background Agilent 6400 LC/MS systems use an Autotune algorithm to optimize the sensitivity and mass resolution of the instrument. Sometimes when running the autotune procedure the autotune does not complete and the user is given an Autotune failed message in MassHunter. One of the main reasons for Autotune failure is a lack of sensitivity or abundance of the tuning ions...

How to Reset the Agilent GC/MSD Tune File to Default Values

Answer

This Information Applies To: Agilent GC/MS MassHunter Acquisition Software

Issue
 
If there is a problem with the current tuning parameters and a need to tune from the default parameters, see the following procedure to reset the EI tune file (atune.u) to the default values.

Steps to follow
Method 1  
  1. Locate the current tune file in Windows File Explorer under the path x:\Masshunter\GCMS\n\5977 (x is the installation drive of MassHunter folder, n is the instrument number). Rename the atune.u tune file ( e.g. atune_old.u ).

  2. Locate and Copy the default tune file (atune.u) using Windows File Explorer from path C:\GCMS\MSSETUP\5977, and paste it into the same location as the renamed tune file in Step 1.

  3. The copied default tune file needs the MS polarity set in Manual Tune view (View>Tune and Vacuum Control > Parameters > Manual Tune). Save the tune file with the correct polarity for your instrument. The polarity information will be in the previous tune reports but also confirm it on the instrument.

    In Figure 1, see the “X” symbol on the right side of the “Neg” box on the MS analyzer label – it means that negative polarity is to be set for this instrument.

     Figure 1. MS Analyzer label – Negative Polarity

  4. Tune the MS before data acquisition after resetting the EI tune file (atune.u) to the default values.

Method 2
Updating or refreshing the instrument configuration in the GC/MS Configuration editor will reset the EI tune file (atune.u) to the default values.
  1. Close the GC/MS Acquisition software and open the Agilent GC/MS Configuration editor.
  2. Click the instrument number on the left at the top to open the instrument configuration.
  3. Do not change any information, just click Ok, when the dialog box opens "proceed with update to instrument configuration," click Yes.
  4. The update is now done, close the Agilent GC/MS Configuration editor.
  5. Tune the MS before data acquisition after resetting the EI tune file (atune.u) to the default values.
The following video (Figure 2.) demonstrates how to update GC/MSD configuration as described in Method 2.

Figure 2. Updating the GC/MSD configuration
 
 
Tip
 
Learn how to effectively operate your Agilent GC/MS MassHunter software:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning course available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help