LC Articles "Could not open channel RD" error message in Lab Advisor
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • +LC Pumps Troubleshooting
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • +LC Detectors Troubleshooting
    • +LC Best Practices
    • +LC Firmware

"Could not open channel RD" error message in Lab Advisor

Created by Sarah Scheirs Sarah Scheirs over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 10 months ago

This Information Applies To: All Agilent liquid chromatography instruments, Lab Advisor


Issue:

The error message 'Could not open channel RD' pops up when performing diagnostic tests in Lab Advisor. Retrying the test is not helping.


Resolution: 

This message appears when another controller is still connected to the module. This could be a data system like OpenLab CDS 2.X, OpenLab CDS ChemStation, OpenLab CDS EZChrom, Empower, or another diagnostic software currently running.

Close the application that is running and try to rerun the tests.

If it is certain that no other software is still connected (be aware that sometimes they run in the background), restart both the PC and the instrument.

After restarting, connect first with Lab Advisor and retry the tests before starting the data system.

 Tip 
Learn how to effectively operate your Agilent Lab Advisor Software: 
Lab Advisor User Manual User manual in Agilent.com

  • Troubleshooting
  • Instrument Control Software
  • Lab Advisor
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help