This Information Applies To: Agilent Infinity II Sample Cooler G7167-60005 and Agilent InfinityLab sample thermostat G7167-60101. These coolers and thermostats are attached on Agilent G7129A/B/C vialsampler, Agilent G7157A Prep Autosampler, and Agilent G7167A/B, G7137A, G5668A, G4767A Multisampler.
Issue
This article outlines the step-by-step instruction and additional information necessary to troubleshoot the cooling related issue, or to provide the Agilent Service Representative with the necessary information to do so.
Background
The integrated sample cooler or thermostat is available as option or upgrade, providing cooling and heating in the range from 4 to 40 °C.
The Agilent Infinity II Sample Cooler is not available anymore and it has been replaced by the Agilent InfinityLab sample thermostat.
Occasionally, especially in the summer, you will face cooling-related problems. For example, the cooler or thermostat does not cool down or drop to an expected temperature, or it throws the following hardware error messages:
|
Note: It takes 45 minutes or more to cool down the sampler. This slow ramping behavior is necessary to avoid icing inside the Sample Cooler/Thermostat.
|
Steps to Follow
Caution: Do not tilt the sampler if there is still some condensate left in the Sample Cooler/Thermostat. It can result in damaging the evaporator fan.
|
Best Practices