ICP-OES Articles How to Reset the AVS 6/7 on Agilent ICP-OES Instruments
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • -ICP-OES Operation
      • ICP-OES Interlocks and Internal Monitoring
      • Water Quality Recommendations for ICP-OES Samples
      • 5000 Series ICP-OES Status Indicators
      • Exhaust Requirements for the 5000 Series ICP-OES
      • Preparing the Instrument Calibration Solution for the Agilent ICP-OES System
      • How to set up communication between Agilent 5000 series ICP-OES Instruments and Agilent ICP Expert software
      • How to perform a Torch Alignment Calibration on the Agilent 700 Series ICP-OES
      • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments
      • SPS4 Alignment and Calibration Wizard
      • How to Upgrade ICP Expert version 7.5
      • ICP-OES Detector and Instrument Calibration
    • +ICP-OES Maintenance
    • +ICP-OES Troubleshooting

How to Reset the AVS 6/7 on Agilent ICP-OES Instruments

Created by Whitney Blackmon Whitney Blackmon over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 8 months ago

This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900), ICP Expert


Issue

How to Reset the AVS 6/7 after maintenance or when a red LED is visible


Steps to follow

 Tip 
Tips: Instructions for changing the AVS switching valve can be found in the ICP Expert Help under “Hardware How to” and “Replace the AVS 4/6/7 Tubing”.


 

Figure 1.  ICP Expert Help

  • After replacing the AVS valve and tubing, a red LED is showing on the AVS plate. Figure 2 shows the location of the AVS LED indicator

Figure 2: AVS plate showing LED indicator

  • Select the Reset function on the Instrument page in ICP Expert (Figure 3).

 

Figure 3: AVS control section on the Instrument Page

 Success  Confirmation: If the AVS LED indicator turns green, the reset was successful and no further actions are needed.
  • If the LED remains red or an error message appears, perform a system reset by power cycling the ICP-OES system by following these steps:
    1. Save and close the software
    2. Turn off the PC
    3. Turn off chiller (if on)
    4. Turn off the autosampler
    5. Turn off the instrument by pressing the front button on the bottom left (while facing the instrument) (Figure 4)
    6. Turn off the main power switch on the left side of the instrument (Figure 5) and wait 5 minutes

  

Figure 4: Front panel power button

 

Figure 5: Main power switch

  •  Power on the ICP-OES system:
    1. Press the main power switch on the instrument to turn the power back on (Figure 5)
    2. Press the front button on the bottom left of the instrument (Figure 4)
    3. Let the instrument fully initialize. The instrument LED indicator should be a light yellow or amber color once this process is completed
    4. Turn on the autosampler and let it fully initialize (the probe will move to all four corners and return to the rinse position)
    5. Turn on the PC
    6. Open the software
    7. Ensure that the instrument is connected to the software (Software will say “Connected” on the top-left corner of the Instrument page and the LED indicator on instrument should be green
    8. Select the Instrument page in the software menu
    9. Is the LED on the AVS plate still red? Try resetting via the software again (Fig. 2)
    10. LED on the AVS plate should be green indicating it is ready to use
    11. Chiller can be turned on and the instrument prepared for use
 Caution 
Caution:  If the AVS LED indicator is still red after this process, please verify that installation was performed properly.  If the red LED persists after reinstallation of the AVS valve and tubing and following the preceding steps, please contact Agilent Technical Support or the local Agilent Representative

  • Operation
  • ICP-OES Accessories
  • AVS 6/7
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help