This Information Applies To: Agilent 700 Series ICP-OES instruments
Issue:
It is important that a torch alignment is performed to ensure that the optics are optimized and are focused on the analytical zone of the plasma. This alignment procedure will ensure maximum sensitivity and signal-to-noise...
This Information Applies To: Agilent 5800/5900 ICP-OES
Issue:
The instrument requires clean, dry, noncorrosive air for cooling purposes. The air enters the instrument through an air supply vent located beside the chimney on the top of the instrument. The vent is fitted with a dust filter to screen...
This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES systems
Issue
The ICP-OES spray chamber filters the sample droplets created by the nebulizer. Larger droplets fall to the chamber drain, while the smaller droplets are introduced to the gas flow going into the ICP torch plasma...
This Information Applies To: Agilent 5000 ICP-OES Series and Agilent ICP Expert Software
Issue
How to connect Agilent 5000 series ICP-OES systems to Agilent ICP Expert software via LAN
Steps to Follow
How to connect Agilent 5000 series ICP-OES instruments to Agilent ICP-Expert software using...
This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900), ICP Expert
Issue
How to Reset the AVS 6/7 after maintenance or when a red LED is visible
Steps to follow
Tips : Instructions for changing the AVS switching valve can be found in the ICP Expert Help under...