This Information Applies To: Agilent Spectroscopy Systems, SPS 4 Autosampler
Issue: When the Agilent SPS 4 autosampler powers up, it performs an initialization routine, which moves the probe arm to predefined positions on all its axes (X, Y, and Theta (rotational). See Figure 1 . If the probe arm...
This Information Applies To: Agilent 5000 Series ICP-OES Systems, AVS 6/7
Issue
Over time the AVS 6/7 rotor can become worn or damaged by chemicals. Agilent recommends inspecting and/or replacing the rotor every 5000 AVS switches or when analytical results or precision fall outside expected limits...
This Information Applies To: Agilent ICP-OES 5100, 5110, 5800, 5900
Issue:
Over time the water inlet filter can become clogged and reduce water flow. Reduced water flow limits the system cooling and can result in the instrument shutting down during analysis.
Recommended maintenance...
This Information Applies To: Agilent ICP-OES Systems
Issue:
The glass concentric Seaspray nebulizer comes by default with the 5000 Series ICP-OES instruments. If running samples which contain high % total dissolved solids (TDS), it is important to first centrifuge or filter them through a 0.45...
This Information Applies To: Agilent 5000 Series ICP-OES, AVS 6/7
Issue
This article shows the required t ubing connections and part numbers between the AVS 6/7 and ICP-OES
ICP-OES sample introduction system:
Figure 1. Sample introduction system and AVS 7 (shown as 1)
Figure 2. Connections...