Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Cleaning a Blocked Nebulizer in the Agilent ICP-OES System
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES
  • Task :  Maintenance Troubleshooting
Related Articles
  • Water Quality Recommendations for ICP-OES Samples

    This Information Applies To: Agilent 5000 series ICP-OES Systems (5100/5110/5800/5900) Issue Discussion of Water Quality for use in the make up of Samples, Blanks, and Standards for ICP-OES Resolution Summary of Recommendations for Water Quality in samples, standards, and blanks for ICP-OES system applications: Low ionic content / resistivity 18 MΩ•cm Low particle content / particles sizes < 0.2 µm Low organic content / Total organic carbon (TOC) Low bacteria content ...
  • Checking for a Blocked ICP-OES Nebulizer

    This Information Applies To: Agilent 5000 Series ICP-OES Issue Blocked nebulizers are the source of many instrument performance issues including loss of sensitivity and poor precision, especially when running organics, high solids, or high salt concentration solutions. The ICP-OES backpressure Nebulizer test verifies nebulizer performance. Run this test if you notice no mist in the spray chamber, have low sensitivity during analysis, or high % RSDs on your replicates. Blockages and partial...
  • Cleaning the ICP-OES Nebulizer

    This Information Applies To: Agilent 500 Series ICP-OES (5100, 5110, 5800, 5900) Issue The ICP-OES nebulizer converts the liquid sample into an aerosol as it enters the spray chamber. A dirty or partially blocked nebulizer traps sample, which leads to poor reproducibility and reduced sensitivity. Droplets on the internal surface of the spray chamber can indicate poor conversion of liquid to mist. ICP-OES overview The ICP-OES sample introduction components are detailed in Figure 1 . ...
  • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735) Issue Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards , memory effects , and the sample introduction system . Sample introduction components include: Autosampler probe Switching valve Pump tubing Nebulizer...
  • Bypassing the AVS for the 5000 Series ICP-OES

    This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7 Issue Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES Tip : These steps should be completed after the Subsystem Communication, Air Flow, Water Flow, Gas Flows, RF Generator, Camera, and Optics Tests have been completed with a status of ‘Done’ or ‘Pass ’ Bypass (or skip) the AVS to run a worksheet and troubleshoot instrument performance issues such as sensitivity...

Cleaning a Blocked Nebulizer in the Agilent ICP-OES System

Answer

This Information Applies To: Agilent ICP-OES Systems


Issue:

The glass concentric Seaspray nebulizer comes by default with the 5000 Series ICP-OES instruments. If running samples which contain high % total dissolved solids (TDS), it is important to first centrifuge or filter them through a 0.45 micron filter prior to running on the ICP-OES. If samples are not fully digested, concentric nebulizers can easily become blocked (partially or fully) by sample particles. Once you have performed the Nebulizer test, as detailed in ICP-OES: Checking for a Blocked Nebulizer, and have learned that your nebulizer may be blocked, this procedure details how to clean the nebulizer.


Steps to follow:

Before removing the nebulizer from the spray chamber, rinse with deionized water. Always wear appropriate safety equipment and follow all regulations for chemical use and disposal.

  1. Turn off the plasma and peristaltic pump.

  2. Remove the sample inlet line tubing (see Figure 1)

  3.  Remove the gas inlet fitting from the nebulizer side arm (see Figure 1). Not all nebulizers are the same. Fittings will vary depending on the nebulizer.



    Figure 1. Nebulizer
    1. Sample inlet tubing, 2. Gas inlet fitting

     Caution 
    Caution: The concentric glass nebulizer is fragile and easily broken. Handle it with extreme care and ensure that the tip is protected from damage. Wear appropriate gloves and clothing.

     

  4.  Unscrew the white knurled nut, or gently twist the nebulizer depending on the nebulizer and spray chamber used to remove it from the spray chamber, being careful not to touch the tip of the nebulizer. See Figure 2


    Figure 2.
    Spray Chamber 
    1. Torch loader handle, 2. Torch, 3. Torch Clamp, 4. Nebulizer gas connections, 5. Spray chamber drain tubing, 6. Spray chamber, 7. Nebulizer, 8. Nebulizer sample tubing
  5. Using a syringe or nebulizer cleaning tool, push water, methanol or rinse solution through the nebulizer in the reverse direction to normal flow. Either push liquid through using the nebulizer cleaning tool or pull it through with a syringe by placing the tip of the nebulizer in the solution and applying a vacuum to the sample inlet end. Rinse with deionized water. 

  6. If the blockage remains, push 25% detergent solution through the nebulizer with a nebulizer cleaning tool and then soak the nebulizer overnight. Rinse with deionized water.

     Warning 
    Warning: Nitric acid is very corrosive and can cause severe burns when it comes into contact with thebody. It is essential that appropriate protective clothing be worn at all times when handling acids. If acid contacts the body, wash off with copious amounts of water and seek medical attention immediately.
  7. If the blockage is still present, soak the nebulizer overnight in concentrated nitric acid. Ensure that no air is trapped in the central capillary and that the nebulizer is completely submerged in the acid.
    Before placing the nebulizer in concentrated nitric acid, use a Pasteur pipette to gently fill the central capillary with acid through the sample inlet of the nebulizer.

  8. Rinse well with water and allow to dry before use.

  9. Reconnect the gas tubing and the sample tubing to the nebulizer.

 Tip 
Learn how to effectively troubleshoot your Agilent ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help