Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Cleaning the 5100/5110 ICP-OES Air Filter
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES
  • Task :  Maintenance
Related Articles
  • Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES

    This Information Applies To: Agilent 5800/5900 ICP-OES Issue: The instrument requires clean, dry, noncorrosive air for cooling purposes. The air enters the instrument through an air supply vent located beside the chimney on the top of the instrument. The vent is fitted with a dust filter to screen out particulate matter from the surrounding environment. This filter should be replaced regularly to prevent reduced air flow. Remove the filter: Figure 1 shows how to remove the air filter...
  • Calculating the Headspace Sampler GC Cycle Time

    This Information Applies To: Agilent headspace samplers (HS) Issue GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause reduced sample throughput as vials wait longer than needed before processing. Background GC cycle time is the total time required for the GC (or GC/MS) system to return to a ready state after a sample run. HS uses the GC cycle time to calculate...
  • Troubleshooting AriaMx Temperature Related Errors at the Start Up After Delivery

    This Information Applies To: Agilent AriaMx real-time PCR system Issue AriaMx showing a series of errors after delivery, guide for troubleshooting and resolving temperature-related error. Background When an AriaMx is delivered, it could take few hours for its temperature equilibrate to room temperature. If the instrument is turned on before it has adjusted to room temperature, it could produce a whole series of thermal related errors ( Figure 1 ). Figure 1 : Errors displayed at the...
  • Routine Maintenance 700 Series ICP-OES Systems

    This Information Applies To: Agilent 700 Series ICP-OES Systems Issue The 700 Series ICP-OES requires regular maintenance, and routine maintenance tasks are able to be performed by the instrument operator. Any maintenance procedures not specifically mentioned in this article should only be performed by Agilent-trained, Agilent-qualified, or Agilent-authorized Field Service Engineers. Warning : Eye Hazard Danger to eyes. The plasma is an intense light source...
  • General Procedure and Common Issues with Licensing of Agilent OpenLab CDS

    This Information Applies To: Agilent OpenLab CDS Software (OpenLab CDS 2.X, OpenLab CDS ChemStation Edition C.01.08 - C.01.10 and OpenLab CDS EZChrom Edition A.05.08 - A.04.10). Issue: All OpenLab CDS software versions need a valid license file (.lic) to work. The creation and insertion of this license file into your software is typically part of the software installation. Nevertheless, you will have to update your license file when: Updating the software or hardware enablements of your license...

Cleaning the 5100/5110 ICP-OES Air Filter

Answer

This Information Applies To: Agilent 5100/5110 ICP-OES


Issue

The instrument requires clean, dry, noncorrosive air for cooling purposes. The air enters the instrument through an air supply vent located beside the chimney on the top of the instrument. The vent is fitted with a dust filter to screen out particulate matter from the surrounding environment. This filter should be cleaned regularly to prevent reduced air flow.


Steps to follow

 Tip 
Tips: This article applies to the standard air filter and does not include the Dusty Environment Filter accessory. 
  1. Turn off the ICP-OES instrument.
     Caution 
    Caution: Always turn off your instrument before attempting to replace a filter to prevent dust and contaminants entering the instrument.
  2. Pull up the air filter handle to remove the air filter.

  3. Vacuum the filter to remove any build-up of dust and debris.

  4. If the filter has an oily build-up, wash the filter in a mild detergent and allow it to dry.

  5. Replace the filter if it is damaged or blocked. Agilent part number G8000-68002.

  6. To install the filter, fit the rounded cut-out over the tab on the air filter housing and then press down while you slide it into place. Purge the instrument for 60 minutes after installing the air filter.

Before using the instrument:

  1. If the external exhaust was turned off, turn on. 

  2. Turn on the instrument and gas supply.

  3. Purge the instrument for at least 60 minutes.

 Tip 
Recommended maintenance: Agilent recommends cleaning the air filter monthly.

 

 Tip 
Learn how to effectively operate your ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education

 

 