Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Replacing and Adjusting the 5000 Series ICP-OES Peristaltic Pump Tubing
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES
  • Task :  Maintenance Operation
Related Articles
  • Replacing and Adjusting the Agilent ICP-MS System Peristaltic Pump Tubing

    This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems Issue Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened or does not require stretching to fit over the pump rollers, it should be replaced with new pump tubing. Background Agilent pump tubing, PVC tubing and connector barb sizes vary with application. Some sized tubing will not require a connector...
  • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735) Issue Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards , memory effects , and the sample introduction system . Sample introduction components include: Autosampler probe Switching valve Pump tubing Nebulizer...
  • Bypassing the AVS for the 5000 Series ICP-OES

    This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7 Issue Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES Tip : These steps should be completed after the Subsystem Communication, Air Flow, Water Flow, Gas Flows, RF Generator, Camera, and Optics Tests have been completed with a status of ‘Done’ or ‘Pass ’ Bypass (or skip) the AVS to run a worksheet and troubleshoot instrument performance issues such as sensitivity...
  • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS

    This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7, ICP Expert version 7.X Issue The AVS can contribute to instrument performance issues such as sensitivity and precision. During troubleshooting, bypassing (or skipping) the AVS and running a worksheet during can identify if the AVS is the root cause of the issue. Other root causes could be the method application, or an instrument fault. Items Required If an internal standard line is required, install the...
  • VGA Maintenance and Cleaning

    Multiple Results
    This Information Applies To: Agilent VGA 77 Accessory Issue: The Agilent Vapor Generation Accessories use high concentrations of HCl and require thorough cleaning to maintain optimum working condition. Warning : The Vapor Generation...
    • Found In:
    • Articles
    • ,
    • Articles

Replacing and Adjusting the 5000 Series ICP-OES Peristaltic Pump Tubing

Answer

This Information Applies To: Agilent 5000 Series ICP-OES (5100, 5110, 5800, 5900) 


Issue

Pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened or does not require stretching to fit over the pump rollers, it should be replaced with new pump tubing.


Peristaltic pump components

Figure 1 shows the peristaltic pump components.

Figure 1:  Peristaltic pump components

1. Pump rollers 2. Pressure bar 3. Pressure bar tensioner screw 4. Pressure bar tensioner 5. Spray chamber to waste 6. Sample to nebulizer

 

 Warning 
Warning: The peristaltic pump has moving parts that can damage hands or fingers caught underneath the clamps during operation. Keep hands away from the peristaltic pump while it is running.

Steps to follow

To install peristaltic pump tubing: 

  1. Install drain pump tubing.
  2. Install sampling pump tubing.
  3. Adjust drain flow rate.
  4. Adjust sampling flow rate.
 Tip 
Tip: Agilent pump tubing, PVC tubing and connector barb sizes vary with application. Some sized tubing will not require a connector barb. Always ensure that the drain tubing has a greater internal diameter than the sample tubing.
To reduce the dead volume and maximize sample throughput, cut and remove excess pump and PVC tubing. Leave at least 10 mm outside the tags on ends of the pump tubing. Use the Agilent tube cutter (part number 8710-19300) for a clean cut with no damage to the tube end (Figure 2). Damage may cause flow rate problems. 

 

Figure 2: Agilent tube cutter

 Install drain pump tubing

Figure 3 shows how to connect drain pump tube to the spray chamber and waste.

 

Figure 3: Connecting the drain pump tubing

 

  1. Attach a piece of PVC capillary tubing to the spray chamber drain.
  2. Attach a connector barb to each end of a drain pump tube. 
  3. Wrap the drain pump tube around the rollers and place the tube tabs into the top and bottom tube retainer positions. 
  4. Connect the pump tube in the bottom retainer position to the spray chamber drain capillary tube. 
  5. Connect the pump tube in the top retainer tube position to a length of tubing that runs to a waste collection vessel. 


 Install sampling pump tubing

Figure 4 shows how to connect the sampling pump tubing to the nebulizer and sample delivery tube.

 

Figure 4: Connecting the sampling pump tubing

  1. Wrap the sampling pump tube around the rollers and place the tube tabs into the top and bottom tube retainer positions. Note: Barbs adaptors are not normally required for sampling pump tubing. The sampling and the nebulizer capillary tubing will insert directly into the sampling pump tubing. 
  2. Connect the tube in the top retainer position to the nebulizer capillary tubing.
  3. Connect the tube in the bottom retainer position to a length of capillary tubing long enough to sit in the sampling solution. Alternatively, if you are using an autosampler, attach the capillary to the autosampler tubing. 
  4. Place the capillary tubing in the solution for uptake. 

Adjust drain flow rate

Figure 5 shows how to adjust the drain pump flow rate.

 

Figure 5: Adjusting the drain flow rate

  1. Turn on the pump.
  2. Turn the pressure bar tensioner screw counterclockwise to release the pressure.
  3. Push the pressure bar towards the drain tubing.
  4. Push the pressure bar tensioner down until the peg at the end of the pressure tensioner sits firmly against the pressure bar.
  5. Gradually tighten the pressure tensioner screw clockwise until the flow through the drain tubing is smooth.
     Tip 
    Tip: To help see the flow, introduce some air bubbles into the capillary by lifting the sample tubing in and out of the sample bottle.

Adjust sampling flow rates:

Figure 6 shows how to adjust the sampling pump flow rate.

 

Figure 6: Adjusting the sampling flow rate

 

 

  1. Disconnect the sample capillary (6 in Figure 1) from the nebulizer and place the end into a suitable container to catch the waste liquid.
  2. Turn on the pump.
  3. Turn the pressure bar tensioner screw counterclockwise to release the pressure.
  4. Push the pressure bar towards the drain tubing.
  5. Push the pressure bar tensioner down until the peg at the end of the pressure tensioner sits firmly against the pressure bar.
  6. Gradually tighten the pressure tensioner screw clockwise until the flow through the sampling tubing is smooth
     Tip 
    Tip: To help see the flow, introduce some air bubbles into the capillary by lifting the sample tubing in and out of the sample bottle.

 

 Tip 
Recommended maintenance: Agilent recommends daily inspection and weekly replacement of pump tubing. 

 

 Tip
Learn more on how to effectively maintain your ICPOES: 
ICPOES-5110-1200e - Agilent 5110 ICP-OES Instrumentation Overview
ICPOES-5900-1200e - Agilent 5800/5900 ICP-OES with ICP Expert Instrumentation Overview e-learning course available from Agilent education

 

 