This Information Applies To: Agilent 5000 Series ICP-OES (5100, 5110, 5800, 5900)
Issue
Pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened or does not require stretching to fit over the pump rollers, it should be replaced with new pump tubing.
Peristaltic pump components
Figure 1 shows the peristaltic pump components.
Figure 1: Peristaltic pump components
1. Pump rollers 2. Pressure bar 3. Pressure bar tensioner screw 4. Pressure bar tensioner 5. Spray chamber to waste 6. Sample to nebulizer
Warning: The peristaltic pump has moving parts that can damage hands or fingers caught underneath the clamps during operation. Keep hands away from the peristaltic pump while it is running.
|
Steps to follow
To install peristaltic pump tubing:
Tip: Agilent pump tubing, PVC tubing and connector barb sizes vary with application. Some sized tubing will not require a connector barb. Always ensure that the drain tubing has a greater internal diameter than the sample tubing.
To reduce the dead volume and maximize sample throughput, cut and remove excess pump and PVC tubing. Leave at least 10 mm outside the tags on ends of the pump tubing. Use the Agilent tube cutter (part number 8710-19300) for a clean cut with no damage to the tube end (Figure 2). Damage may cause flow rate problems.
|
Figure 2: Agilent tube cutter
Figure 3 shows how to connect drain pump tube to the spray chamber and waste.
Figure 3: Connecting the drain pump tubing
Figure 4 shows how to connect the sampling pump tubing to the nebulizer and sample delivery tube.
Figure 4: Connecting the sampling pump tubing
Figure 5 shows how to adjust the drain pump flow rate.
Figure 5: Adjusting the drain flow rate
Tip: To help see the flow, introduce some air bubbles into the capillary by lifting the sample tubing in and out of the sample bottle.
|
Figure 6 shows how to adjust the sampling pump flow rate.
Figure 6: Adjusting the sampling flow rate
Tip: To help see the flow, introduce some air bubbles into the capillary by lifting the sample tubing in and out of the sample bottle.
|
Recommended maintenance: Agilent recommends daily inspection and weekly replacement of pump tubing.
|
Learn more on how to effectively maintain your ICPOES:
ICPOES-5110-1200e - Agilent 5110 ICP-OES Instrumentation Overview
ICPOES-5900-1200e - Agilent 5800/5900 ICP-OES with ICP Expert Instrumentation Overview e-learning course available from Agilent education |