This Information Applies To: Agilent ICP-MS MassHunter Software 4.x (Rev C)
Issue
It is best practice to note MassHunter Software Settings before running the Glitch Resetter, renaming MassHunter configuration settings files, or before upgrading/reloading the ICP-MS MassHunter software.
...
This Information Applies To: Agilent ICP-MS MassHunter Software revisions 4.x and 5.x
Issue
How to clear or archive the acquisition queue? How to resolve queue task file errors?
Background
The Acquisition Queue ( Figure 1 ) lists all tasks completed by the instrument in the ICP-MS MassHunter...
This Information Applies To: Agilent ICP-MS MassHunter 5.1 Software, SPS 4 Autosampler
Issue
The SPS-4 autosampler (ALS) may need to be reconfigured after the Glitch Resetter tool is used to reset the software or after a change from a different autosampler, such as the ASX-520. This process will...
This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
Elevated doubly charged values that are greater than 3% for x-lens configuration or greater than 7% for s-lens configuration, can be caused by contamination, or an issue with the sample introduction system.
Background...
This Information Applies To: Agilent ICP-MS Systems
Issue
The Helium cell gas line on the ICP-MS may be contaminated with air and moisture in case of a tank replacement, or if the instrument was not used for an extended amount of time. Purging the Helium cell gas line helps ensure a steady supply...