Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-MS Portal
  • Agilent Knowledge Portal
ICP-MS Portal
Articles Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-MS Portal Home
  • ICP-MS Portal Articles
  • ICP-MS Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Spectroscopy Accessories ICP-MS Systems ICP OES Systems
  • Component :  ICP-MS Single Quadrupole ICP MS Triple Quadrupole
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES 700 Series ICP-OES 7800 ICP-MS 7850 ICP-MS 7900 ICP-MS 8800 Triple Quadrupole ICP-MS 8900 Triple Quadrupole ICP-MS Recirculating Chiller
  • Task :  Troubleshooting
Related Articles
  • How to Power On Agilent ICP-MS Systems

    This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900) Issue Powering on Agilent ICP-MS Systems after maintenance, long-term shutdown, or for power cycling. Steps to Follow Confirm that the ICP-MS power plug is plugged in to the ICP-MS and to the wall. Press the main power breaker switch on the back of the ICP-MS to the ON (I) position ( Figure 1 Item 2). Figure 1 . ICP-MS Rear View 1. Foreline pump breaker switch, 2. Main power breaker switch ...
  • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments

    This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900), ICP Expert Issue How to Reset the AVS 6/7 after maintenance or when a red LED is visible Steps to follow Tips : Instructions for changing the AVS switching valve can be found in the ICP Expert Help under “Hardware How to” and “Replace the AVS 4/6/7 Tubing”. Figure 1. ICP Expert Help After replacing the AVS valve and tubing, a red LED is showing on the AVS plate. Figure 2 shows the location...
  • ICP-OES Detector and Instrument Calibration

    This Information Applies To: Agilent 5000 Series ICP-OES (5100, 5110, 5800, 5900) Issue How to calibrate the ICP-OES ICP-OES calibration is a two-step procedure: The detector calibration performs a dark current calibration to correct for detector and electronic background signals. The instrument calibration uses a specified calibration solution to calibrate the wavelengths of the polychromator. ICP-OES wavelength calibration solution is available as either Agilent ICP-OES wavelength...
  • Calibration Failure Troubleshooting on ICP-OES Systems

    This Information Applies To: Agilent 5000 series ICP-OES systems Issue Instrument calibration (wavelength), detector calibration (dark current, or method calibration (worksheet) failure. Resolution There are several reasons why a calibration may fail. The following is a list of common areas to check if problems are encountered with calibrations. The steps to follow will depend upon whether the calibration tests start, whether all the lines in your calibration have failed, or whether...
  • Cleaning the 5000 Series Agilent ICP-OES Water Inlet Filter

    This Information Applies To: Agilent ICP-OES 5100, 5110, 5800, 5900 Issue: Over time the water inlet filter can become clogged and reduce water flow. Reduced water flow limits the system cooling and can result in the instrument shutting down during analysis. Recommended maintenance schedule : Agilent recommends cleaning the water inlet filter every three months. ICP-OES water inlet filter overview: Figure 1: ICP-OES water inlet filter components 1 Water outlet connected...

Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Answer

This Information Applies To: Agilent G3292A and G8481A Chillers used on ICP-MS and ICP-OES instruments


Issue 

When experiencing issues with your Agilent chiller, please try a factory reset first, as shown in the following steps and in Figure 1.


Resolution 

1. Place the main power switch on the rear of the instrument in the off position.

2. Press and hold the soft power button on the front panel while returning the main power switch to the on position (O).


Figure 1. Performing a Factory Reset of the Chiller

After performing a factory reset, you will need to set the chiller temperature to 20 °C:

1. Press the Select/Set knob on the front panel (right above the soft power button).

2. The decimal point at the bottom right of the display will flash, indicating that the temperature can be changed.

3. Rotate the Select/Set knob until the desired setpoint temperature (20 °C) is displayed.

4. The setting is accepted after the Select/Set knob is pressed a second time or automatically after a few seconds of inactivity.

If the factory reset doesn't help, and you are getting a "Lo H2O" error on the front display of the chiller, please check the fuses after ensuring that the chiller is full of coolant:

 Warning 
Warning: Always unplug the chiller before removing any covers.

 

1. Unplug the chiller.

2. From the back of the chiller, slide off the top cover after removing the two screws on the left and right side, as shown in Figure 2 and Figure 3:


Figure 2. Remove the screws at the top on the back of the chiller.


Figure 3. Remove the top cover of the chiller.

3. Then remove the two screws at the top of the front of the chiller, as shown in Figure 4.


Figure 4. Remove the screws at the top of the front of the chiller.

4. Once you have access to the main board, identify the two fuses, as shown in Figure 5.

 
Figure 5. The two fuses on the main board of the chiller.

5. Remove and check the continuity of the two fuses with a multi-meter.

6. If new fuses are needed, they can be obtained at any home improvement store.

7. Replace the bad fuses, and reassemble the screws and the top cover.

8. Turn on the main power to the chiller.

 

 Tip 
Learn how to effectively troubleshoot your ICP-OES instrument:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education

 

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help