This Information Applies To: ICP-MS MassHunter 4.x/5.x
Issue
Reviewing the latest Performance Report, and Performance Report History is a useful tool for troubleshooting ICP-MS instrument performance issues. Historical sensitivity, background, precision, hardware settings, tune parameters, and meter readings can be reviewed to compare what has changed since the last “good” performance report.
Steps to Follow
The performance report can be exported as a *.PDF or as a *.xml file. If you are troubleshooting with Agilent Technical Support, you might be instructed to email either the *.xml or *.PDF file.
Click here for instructions to export the PerformanceReport.xml file
Click here for instructions to export to PDF for MassHunter 5.x
Click here for instructions to export to PDF for MassHunter 4.x
Export the PerformanceReport.xml File
Tips: If you see other files in this PerformanceReport folder (often with date and time stamps), they are archived performance report files and do not need to be exported. Only export the PerformanceReport.xml file.
|
Export to PDF for MassHunter 5.x
Export to PDF for MassHunter 4.x
Learn how to effectively troubleshoot your Agilent ICP-MS System: Agilent 7800/7850/7900 ICP-MS with MassHunter Agilent 8900 Triple Quadrupole ICP-MS with MassHunter e-learning courses available from Agilent education |