Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
Articles How to Blank off a GC/MS Instrument for Troubleshooting
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC/MS Portal Home
  • GC/MS Portal Articles
  • GC/MS Portal Videos
  • More
  • Cancel
  • Created :  10 Mar 2023
  • Modified :  17 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC/MS Systems
  • Component :  GC/MS Single Quadrupole GC/MS Triple Quadrupole
  • Product Name :  5973 Series GC/MSD 5975 Series GC/MSD 5977 Series GC/MSD 7000 Series Triple Quadrupole GC/MS 7010 Series Triple Quadrupole GC/MS
  • Task :  Troubleshooting
Related Articles
  • Troubleshooting an FID not Igniting

    This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000 Issue: This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out. Cause: If the FID is having ignition issues, it is due to the following causes: FID temperature : is too low. Lit Offset : is incorrect. FID gas flows : if the air, hydrogen, or makeup gas flows are not correct (or not meeting setpoints). Gas...
  • How to Blank off a GC Detector for Troubleshooting

    This Information Applies To: Agilent 6890/7890/7820/8860/8890 GC systems Issue Blanking off a GC detector can help locate the source of issues such as baseline noise, baseline spikes, elevated baseline, or detector output/performance issues. Items Required For Capillary Columns No-hole ferrule (p/n 5181-3308 ) Finger tight blanking nut (p/n 5020-8294 ) Capillary column nut For Packed Columns Brass female blanking nut 1/8 " (p/n 5180-4124) Brass female blanking nut...
  • How to Run a Single Sample with ALS in Agilent GC/MS MassHunter Data Acquisition Software

    This Information Applies To: Agilent GC/MS MassHunter Data Acquisition Software Issue Preparing and running a sample is essential for data acquisition. The steps below detail how to run a single sample using an Automatic Liquid Sampler (ALS) on your GC/MS system. Steps to follow The following steps to run a single sample using a previously created method are demonstrated in the video shown in Figure 1 . 1. Browse and select the method you wish to use. 2. Select Method > Run Method...
  • Using the Condensation Test to Troubleshoot the GC Inlet

    This Information Applies To: Agilent GC Systems Issue This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler. Background Typical chromatographic features that indicate possible contamination problems include ghost peaks, carryover, excessive baseline noise or baseline drift, and instability. See Basic Troubleshooting for GC Systems . Steps to...
  • Checking the Instrument Status of your Agilent Single Quadrupole LC/MS System Using OpenLab CDS

    This Information Applies To: Agilent Single Quadrupole LC/MS systems controlled by OpenLab CDS software Issue How to check and recognize the current Instrument Status of your 6100 Series and InfinityLab LC/MS systems using OpenLab CDS software. Steps to Follow If necessary, in the top ribbon, click the Home tab, the click Instrument Status. Figure 1 . Instrument Status option location In the Instrument Status window, click the Dashboard tab to see the SQ Instrument Status...

How to Blank off a GC/MS Instrument for Troubleshooting

Answer

This Information Applies To: Agilent single quadrupole and triple quadrupole GC/MS systems


Issue

During troubleshooting, it can be useful to isolate the MS detector from the rest of the GC system (sample introduction, inlet, column etc.).


Background

Blanking off the MS will help to determine if an issue is coming from the MS detector itself or the GC. Some examples where this procedure might be useful:

  • Noisy or unusual baseline.
  • Tuning difficulties.
  • Leaks (high oxygen and nitrogen on the tune report) where the common MS leak troubleshooting was not successful.
  • Large number of peaks in tune report.

Items Required

  • Long no hole ferrule (p/n 5181-3308) paired with column nut for MS interface (p/n 05988-20066)
  • Short no hole ferrule (p/n 5190-4054) paired with Self Tightening column nut (p/n 5190-5233)

Steps to Follow

  1. Load the GC maintenance method. 

  2. Turn off the MS transfer line Aux heater, from the controlling GC/MS software or GC interface (keypad/touch screen)

  3. Vent the GC/MS system.

  4. Once the GC/MS is cool and vented, remove the column from the MS interface. Unscrew the column nut and carefully remove the column. It can be left in the oven, but be careful not to get the column end dirty or scratched. 

     

       Warning: Due to risk of explosion, when using hydrogen carrier, do not leave the column venting into the oven. Either cap the end of the column with a septum, or turn off the inlet pressure and leave the oven under 50 °C degrees Celsius.

     

  5. Install the no-hole ferrule into the MS column nut to create a blanking nut. Remember that the pointed side of the ferrule should face into the nut and the flat side of the ferrule faces towards the MS interface.

  6. Screw the blanking nut onto the MS interface.

    • For the regular column nut, screw on finger tight then us a spanner to tighten the nut an extra 1/4 to 1/2 turn. You will need to tighten slightly more after it has heated up.

    • For the Self Tightening column nut, screw in until it is finger tight (careful not to overtighten)

  7. Pump down the MS.

  8. After sufficient pump down time (and making sure that the new blanking nut is leak free), you can begin testing the MS:

    • Run an autotune. It may look slightly different (there is no column flow into the source) but assess if the issue is resolved.

    • If the issue occurred when an Acquisition Method was run, run a blank injection to see if the issues occur (except if you are using hydrogen carrier gas).

  9. Assess the results;

    • Is the issue still present? Then the problem is located to the MS, focus troubleshooting efforts there.

    • Is the issue no longer there? Then the GC system - column, CFT splitters or fittings, inlet, sample introduction, will likely be the source of the issue. 

 

 
Learn how to effectively operate and Maintain your Agilent GC/MS system: 
Agilent 5977 GC/MSD with MassHunter Workstation
Agilent 5977 GC/MSD with OpenLab CDS
Agilent 7000 or 7010 Triple Quadrupole GC/MS with MassHunter Workstation e-learning paths available from Agilent education

 