This Information Applies To: 5977A and 5977B
Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut.
Warning : a. The GC and MS operate at high temperatures...
This Information Applies To: 7890 or 8890/8860 GCs with a split/splitless inlet, MMI, PTV inlet, or volatile interface inlet.
Issue
Check the split vent trap line for contamination or an abnormal restriction causing high inlet pressure.
Items...
This Information Applies To: Agilent GC Systems
Issue
The GC carrier gas actual pressure exceeds its pressure setpoint.
Resolution
Split/splitless inlet, MMI inlet, PTV inlet, or volatiles interface
Check the split vent trap line for contamination...
This Information Applies To: Agilent 8890/8860 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent 7890/7820 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...