This Information Applies To: Agilent 8890, 8860, 7890, 7820, 6890, 6850 GC systems
Issue
Manual injections are needed for a gas chromatograph without an automated sampler. They are also used for troubleshooting specific issues such as no peaks, lower peaks than expected, ghost peaks, and other chromatographic problems.
Items Required
Steps to Follow
Prepare your syringe, sample, and hardware
Locate a suitable syringe. If you have an ALS (7683 or 7693), you can use the syringe from the injector tower. See Removing and Installing a Syringe on the Agilent 7693A Autosampler.
Prepare a sample or standard.
Verify that the syringe is working.
Manually draw up some sample or solvent (put the end of the syringe in the liquid, then pull up the plunger to draw up the liquid)
Push the plunger down to eject the liquid onto a paper towel.
Verify that you can see the drops exit the syringe. If not, the syringe is clogged.
Prepare the hardware so the top of the inlet is easily accessible. You need to be able to see the green septa nut (Figure 1):
Figure 1. Septa nut for an SSI and MMI
Prepare the method in the software
From the top menu, click Instrument > Select Injection Source... (Figure 2)
Figure 2. Select Injection source...
In the Select Injection Source/Location dialog box, select the manual option, then select the injection location, and click OK (Figure 3).
Save the acquisition method with a name that identifies it as manual injection.
To start a single sample, from the top menu, click Run Control > Run Method (Figure 4)
Perform the manual injection
Learn how to effectively operate your Agilent GC system:
GC-0GEN-1040z - Practical Steps in GC Troubleshooting GC-8890-2200s - Agilent 8890 GC Maintenance and Troubleshooting e-learning courses available from Agilent education |