This Information Applies To: Agilent GC systems (8890, 8860, 7890, 7820, 6890, 6850, 6820) using Agilent OpenLab CDS
Issue
Manual injections are needed for a gas chromatograph without an automated sampler. They are also used for troubleshooting specific issues such as no peaks, lower peaks than expected, ghost peaks, and other chromatographic problems.
Items Required
Steps to Follow
Prepare your syringe, sample, and hardware
Locate a suitable syringe. If you have an ALS (7683 or 7693), you can use the syringe from the injector tower. See Removing and Installing a Syringe on the Agilent 7693A Autosampler.
Prepare a sample or standard.
Verify that the syringe is working.
Manually draw up some sample or solvent (put the end of the syringe in the liquid, then pull up the plunger to draw up the liquid).
Push the plunger down to eject the liquid onto a paper towel.
Verify that you can see the drops exit the syringe. If not, the syringe is clogged.
Figure 1. Septa nut for an SSI and MMI
Prepare the method in the software
For a Sequence
Figure 2. Sequence Injection source
For a Single Sample
Figure 4. Single Sample Analysis
Perform the manual injection