This Information Applies To: Agilent 4200, 4150 and 2200 TapeStation.
Issue
Is it possible to analyze single stranded DNA (ssDNA) on the TapeStation? Which kit is the most suitable for analysis of ssDNA?
Resolution
Agilent does not offer any...
This Information Applies To: Agilent 4200 and 4150 TapeStation systems
Issue
How to Run the software installation verification on 4200 and 4150 TapeStation systems.
Background
The 4150 TapeStation model G2992A and 4200 TapeStation model G2991B...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, and Femto Pulse
Issue
Perform a capillary alignment when a new array is installed in your Agilent Parallel Capillary Electrophoresis instrument. An alignment may also be performed...
This Information Applies To: Agilent 4200 and 4150 TapeStation
Issue
Are the Agilent DNA and RNA ScreenTape assays denaturing?
Resolution
The DNA ScreenTape assays in the TapeStation portfolio are not denaturing. Only the RNA and High Sensitivity...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, ZAG DNA Analyzer, Femto Pulse
Issue
Instructions for replacing the syringe on parallel capillary electrophoresis (PCE) instruments
Steps to Follow
Using a flathead...