Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Atomic Absorption Articles Alignment of the Sample Probe on the Agilent PSD 120
  • Atomic Absorption Articles
  • Atomic Absorption Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Atomic Absorption
    • -Atomic Absorption Operation
      • How to Set up the PSD Probe Arm
      • Alignment of the Sample Probe on the Agilent PSD 120
      • Flame Optimization for Agilent Atomic Absorption Systems
      • AA Calibration Curves Types
      • AA software - SpectrAA Flags
      • VGA Setup
      • Shutting Down the Agilent Vapor Generation Accessory (VGA)
    • +Atomic Absorption Maintenance
    • +Atomic Absorption Troubleshooting
    • +Atomic Absorption Best Practices

Alignment of the Sample Probe on the Agilent PSD 120

Created by Tina Harville Tina Harville over 2 years ago | Last modified by Carlos Vargas Carlos Vargas over 1 year ago

This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System


Issue:

It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120, so that the sample is injected properly into the graphite tube.


The correct alignment of the probe (the end of the capillary protruding from the base of the probe holder) is:

  • The probe locates in the center of the vial when it collects the solution.
  • The probe depth is sufficient to remain under the liquid during the drawing up phase.
  • It enters and exits the rinse solution vessel without touching the sides or base. 
  • Enters and exits the graphite tube without touching the sides of the injection hole.

Steps to follow: 

  1. Make sure that the GTA workhead is properly aligned with the optical path and that your PSD is properly installed.

  2. Place an empty vial on the dispenser carousel in position 1.

  3. Create a method worksheet. See the online Help.

  4. Go to the Analysis page. Select Instrument menu > Furnace Facilities. Turn on the camera (if installed).

  5. Click Align. The autosampler will position the capillary in position 1 in a completely down position. (See Figure 1)


    Figure 1. Graphite Furnace AA Sample Probe Alignment

  6.  Verify that the probe is correctly aligned to the sample vial. The probe should almost reach the bottom of the vial, but not actually touch it (See Figure 2). 



     Figure 2. Probe alignment in vial

  7. If the end of the probe is too far from the bottom of the vial, or if it touches the bottom of the vial:
    a. Loosen the probe guide screw and adjust the length of the capillary as necessary.
    b. If necessary, loosen the screw on the probe holder and adjust the angle of the probe until it is aligned vertically in the vial.

  8. Once the correct length of the probe in the vial has been established, click OK in the confirmation window. The probe will move to the furnace head.

  9. Place a finger under the probe arm as the capillary begins to descend to the oven and gently lower the hand arm.

  10. Check the capillary travels through the entry hole of the furnace tube and does not touch the sides. If necessary, release the locking screw located below the PSD, adjust the position, then relock the screws. 

  11. Lower the probe into the graphite furnace. Use the furnace camera window or the viewing mirror to check the location of the tip inside the tube (See Figure 3). If necessary, use the depth adjustment screw to adjust the tip so it is about 1 mm above the base of the graphite tube. Tighten the locking screws. 


    Figure 3. Lowering the Capillary into the Graphite Furnace

 Tip 
Learn how to effectively operate your Agilent Graphite Furnace AA:
240/280 AA User's Guide (Including Zeeman) user manual on Agilent.com and the video Troubleshooting and Maintenance of Atomic Absorption Systems: Part 3- Graphite Furnace AAS
  • GTA 120
  • AAS Instruments
  • Operation
  • 240FS AA
  • 240Z AA
  • psd 120
  • 280Z AA
  • AAS Accessories
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help