This Information Applies To: Agilent SIPS 10/20 and Flame AA
Issue
Replacing and conditioning the SIPS pump tubing.
Background
Whenever new pump tubes have been installed, it is necessary to condition the tubing before the analysis to prevent signal drift as the tubing settles during operation.
Tips: For recommended maintenance of the SIPS 10/20, see SIPS 10/20 Maintenance.
|
Steps to Follow
Replacing the Pump Tubing
When pump tubing needs to be replaced, it is recommended to follow these additional maintenance and operational procedures:
Tips: For an overview of the use of the SIPS 20 with a Flame AA instrument and SpectrAA Software, see Agilent Fast Sequential (FS) AA Instruments, SIPS 20 and PROMT.
|
Conditioning the SIPS Pump Tubing
Before each use of a new tube, the tubing should be conditioned using the following procedure:
Tips: Triton X-100 is also known as octyl phenol decaethylene glycol ether, CAS No. 9002-93-1
|