This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to optimize the Agilent Flame AA for better instrument performance and response stability
Steps to follow:
Use Figure 1 as reference for the procedure on the Flame AA optimization.
Figure 1. Agilent Flame AA instrument diagram
|
Tips: Steps 1 through 4 are performed with the flame off in the Optimization Page of SpectrAA by selecting Optimize Lamps
|
Figure 1. Lamp Optimization Page
|
Tips:See Troubleshooting and Maintenance of Atomic Absorption Systems: Part 1 - Hollow Cathode Lamps - YouTube for more detailed information
|
|
Tips: See What are the typical % Gain or EHT values for hollow cathode lamps? | Agilent for estimated Gain % values
|
Figure 2. Vertical Alignment of Burner
Figure 3. Horizontal Adjustment Control
Figure 4. Alignment of Lamp light on one side of burner
Figure 5. Alignment of Lamp light across Burner slot
|
Tips: See Troubleshooting and Maintenance of Atomic Absorption Systems: Part 2 - Flame Atomic Absorption - YouTube for more detailed information
|
|
Tips: The following steps are performed with the Flame On in the Optimization Page and select Optimize Signal
|
|
Warning: Once the flame is on, keep loose articles and hands away to avoid serious burns. Burner will be hot.
|
Figure 6. Graduated Cylinder with DI water
Figure 7. Nebulizer Rate adjustment
Figure 8. Optimize Signal and blue signal bar
Figure 9. Glass Bead adjustment
Figure 10. Fuel Adjustment
|
Tips: See Troubleshooting and Maintenance of Atomic Absorption Systems: Part 2 - Flame Atomic Absorption - YouTube for more detailed information
|
|
Confirmation: Optimization is now complete
|
|
Tips: If using SIPS, start the pump tube conditioning program and complete the bubble test to confirm that an air bubble is moving towards the instrument. This procedure verifies that no clogs are present and that the nebulizer uptake is at least 5 mL/min
|