Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Articles Troubleshooting and Maintenance of AA Systems Part 2B- Flame Optimization
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Atomic Absorption Portal Home
  • Atomic Absorption Portal Articles
  • Atomic Absorption Portal Videos
  • More
  • Cancel
  • Created :  23 Mar 2023
  • Modified :  23 Mar 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Absorption Systems
  • Product Name :  240FS AA 280FS AA 55B AA
  • Task :  Maintenance Troubleshooting
Related Articles
  • Troubleshooting and Maintenance of AA Systems: Part 2A- Cleaning Flame AA Components

    This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS Issue: The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber. Disassembling and Cleaning the Spray Chamber Remove the chimney and flame shield. Remove the burner by lifting it upwards and slightly twisting. Warning : Extinguish the flame and ensure that the burner is cool before removing...
  • How to Clean a Blocked Nebulizer of an Agilent Flame Atomic Absorption System

    This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA) Issue: How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min. Steps to follow: Warning : Never remove a nebulizer from the spray chamber while the flame is on. Always extinguish the flame before removing the nebulizer from the spray chamber. Parts in and near the burner compartment become hot during operation...
  • Flame Optimization for Agilent Atomic Absorption Systems

    This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA) Issue: How to optimize the Agilent Flame AA for better instrument performance and response stability Steps to follow: Use Figure 1 as reference for the procedure on the Flame AA optimization. Figure 1. Agilent Flame AA instrument diagram Tips : Steps 1 through 4 are performed with the flame off in the Optimization Page of SpectrAA by selecting Optimize Lamps ...
  • Replacing a Hollow Cathode Lamp

    This Information Applies To: Agilent atomic absorption systems (AA 240/280/55B/240Z/280Z) Issue Replacing hollow cathode lamps is a common maintenance procedure and can be used during troubleshooting to isolate the source of an error. Steps to Follow If you are developing a method using an uncoded lamp, install the lamp as described in the following procedure and note the position. The numbers of each lamp position are clearly marked. Enter the lamp position number for the current method...
  • General Maintenance for the Agilent Flame and Graphite Furnace Atomic Absorption Systems

    This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace) Issue: This article includes recommended maintenance procedures to be performed daily, weekly or annually to ensure optimum performance of the Agilent atomic absorption systems. Steps to follow: Schedule of recommended maintenance procedures: Daily Check that the exhaust system has positive extraction Check the gas supplies Check...

Troubleshooting and Maintenance of AA Systems Part 2B- Flame Optimization

Answer

This Information Applies To: Agilent Flame AA Systems: 55B, 240FS, 280FS 


Issue:

After cleaning and re-installing the flame atomic absorption system components, it is important to optimize the burner position, nebulizer uptake rate, flame conditions, glass impact bead, and mixing paddles. Proper optimization ensures that the atomic absorption system will yield accurate and precise results throughout its lifetime.


Steps to follow:

  1. Optimize the burner position vertically, horizontally, and rotationally using the burner alignment card.

  2. Attach the plastic capillary tubing to the nebulizer. Wind the nebulizer locking thimble and locking ring clockwise, as far as they go.

     Figure 2

  3. To determine the nebulizer uptake rate, ignite the flame, and aspirate distilled water from a small graduated cylinder over a one minute time period. Adjust the thimble until the uptake rate is between 5-6 mL per minute (ml/min), then tighten the locking ring (behind the thimble).

    Figure 3

  4. Select or develop a Cu method to use for optimization.  Go to the worksheet window and click optimize and the hollow cathode will complete the peaking procedure. Once complete, click Optimize Signal > Instrument Zero.


  5. Aspirate a 5 ppm Cu solution. Optimize the burner position vertically, horizontally, and rotationally to yield the highest signal.

  6. Then adjust the glass impact bead position to achieve the maximum signal and minimum noise. Close the optimization window by clicking OK.

  7. For most flame AA applications, the mixing paddles should be fitted in the spray chamber, and the glass impact bead should be adjusted for best sensitivity. If you have more difficult samples to run, such as high % total dissolved solids (TDS), ensure that the mixing paddles are in place inside the spray chamber.  Adjust the impact bead at least half a turn clockwise from the optimum sensitivity position. This will give better performance with difficult sample types.

See the other articles in this series:  Troubleshooting and Maintenance of AA Systems: Part 1-Hollow Cathode Lamps and Troubleshooting and Maintenance of AA Systems: Part 2A- Cleaning Flame AA Components

 

 
Learn how to effectively maintain and troubleshoot your Agilent Atomic Absorption system:
240/280 User's Guide 
Troubleshooting and Maintenance of Atomic Absorption Systems: Part 2 - Flame Atomic Absorption Video on the Agilent Knowledge Portal
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help