This Information Applies To: All Agilent Spectroscopy Systems (AA50,55,240,280)
Issue: The deuterium (D2) lamp will need replacing if the lamp fails to light or it cannot match the hollow cathode element lamp energy in the Optimize page in SpectrAA Software. See Figure 1.
Figure 1. Optimize page in SpectrAA Software
Steps to Follow:
Warning: Hot Surface and Eye Hazard. The deuterium lamp and its cover become very hot during operation. To prevent skin burns, allow the assembly to cool down before removing it.
|
Warning: The deuterium lamp emits dangerous levels of UV radiation. To avoid eye and skin damage, NEVER interfere with the interlock switch, or attempt to operate the lamp unless it is mounted in the instrument. |
Figure 2. Video showing D2 lamp replacement
To change the deuterium lamp, follow these steps. For a video of lamp replacement, see Figure 2.
Figure 3. D2 Lamp with bracket
Tip: After installation, a new deuterium lamp must be aligned. See steps below.
|
To align a deuterium lamp:
Tip: If both lamps are correctly aligned, the 2 scales (HC and D2) on Figure 1, will be approximately the same height. |
Learn how to effectively maintain and troubleshoot your Agilent Spectroscopy System): Guidelines for Troubleshooting and Maintenance of AA Systems
|