Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Molecular Spectroscopy
  • Technical Areas
Molecular Spectroscopy
Wiki Cary UV Workstation Method Development FAQs
  • Announcements
  • Blog
  • Forum
  • Wiki
  • More
  • Cancel
Actions
  • Share
  • More
  • Cancel
Table of Contents
  • -Molecular Spectroscopy Support, Applications & eLearning Resources
    • -UV-Vis & UV-Vis-NIR Spectroscopy Resources
      • -Cary 3500 FAQs
        • Cary 3500 General Operation FAQs
        • Migration from Cary WinUV or UV ChemStation Software Comparison Tool
        • Cary UV Workstation Data Analysis FAQs
        • Cary UV Workstation Method Development FAQs
        • Cary 3500 UV-Vis Sample Temperature Probe FAQs
    • FTIR Spectroscopy Resources
    • Raman Spectroscopy Resources
    • Fluorescence Spectroscopy Resources
    • LDIR Chemical Imaging Spectroscopy Resources
Still Need Help?

Post your question in our User Forum or Contact Support.

Cary UV Workstation Method Development FAQs

Created by Agilent Agilent over 2 years ago | Last modified by kristen.sensabaugh kristen.sensabaugh over 2 years ago

Is there a fast and easy way to move between parameters when setting up a method?

Yes, you can “tab” through parameters on the setup and sequencer pages.

 

Can you change any method parameters once you’ve started or completed a sequence?

No, all parameters are locked once an analysis has started. To change any parameters, a new worksheet must be created. Click “Save As” and select the Method check box. This will enable you to work with the current parameters and edit them before measuring new samples.

 

How do I enter individual wavelength values into a Concentration (or any) method?

Ensure the “Select Mode” has been toggled to “Wavelength”, enter the value then press the Return Key on the keyboard.

 

Can we set more than one measurement setpoint in a kinetics application? I.e.: make measurements at 1 minute intervals for 10 minutes, then 5 minute intervals for another 20 minutes? 

We have not implemented that functionality for this release.

 

Can you restart a method, or re-baseline (say that you forgot to put your reference cell in), without creating an entirely new method?

Yes. If you navigate to the sequence page, you can recheck the baseline entry in the sequence table.

 

Does the method “default” to include a baseline? 

No, the current default setting is to not to include a baseline/zero.

 

More FAQs: Cary 3500 FAQs

 

Keywords: Cary 3500 UV-Vis Spectrophotometer, UV-Vis, simultaneous scanning, temperature controlled UV-Vis, Compact UV-Vis, Compact Peltier UV-Vis, Multicell UV-Vis, Multicell Peltier UV-Vis, Multizone UV-Vis system, Cary UV Workstation,

  • cary 3500
  • cary uv workstation
  • multicell peltier uv-vis
  • compact peltier uv-vis
  • uv-vis
  • UV-Vis and UV-Vis-NIR
  • temperature controlled uv-vis
  • simultaneous scanning
  • cary 3500 uv-vis spectrophotometer
  • multizone uv-vis system
  • multicell uv-vis
  • compact uv-vis
  • faq
  • Share
  • History
  • More
  • Cancel
  • Sign in to reply
  • Agilent
    Agilent over 2 years ago in reply to Adriaan

    Hi Adrian, 

    Thanks for the question,

     

    Re:

    Q: In a concentrations method; what is the use of Collect Mode "Scan"?

    A: ?

     

    The Scan mode is super helpful in the concentration application. This allows users to collect full scans and then come along and slice at appropriate wavelengths to automatically build the corresponding calibration curve at the selected wavelength.

    This is particularly nice when you are working with a material that you are perhaps not very familiar with. Or perhaps just to have the opportunity to interrogate the data more comprehensively, i.e. at different wavelengths.

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • Adriaan
    Adriaan over 3 years ago

    Q: In a concentrations method; what is the use of Collect Mode "Scan"?

    A: ?

     

    Q: In a concentrations method; can custom concentration units be used instead of the preset options?

    A: Yes, you can type a custom concentration unit into the "Units" box. Ensure that you use XXX/YYY format, i.e. mg/kg.

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Privacy Statement
Terms of Use
Contact Us
Site Help