Followed the LAN setup in the manual, yet all I get is Powerfail. Using Chemstation version B.05.04 (which is supposed to work on Windows 10).
Followed the LAN setup in the manual, yet all I get is Powerfail. Using Chemstation version B.05.04 (which is supposed to work on Windows 10).
Hello monty_python66 :
Please see the attached communication troubleshooting guide. One possibility is that you do not have the addresses set to use the default LAN card address. The default address of the G1369x LAN card is 192.168.254.11 with switches 5&6 up. The switches are located on the card. Set your computer to 192.168.254.10.
Best regards,
Mike
Hello monty_python66 :
Please see the attached communication troubleshooting guide. One possibility is that you do not have the addresses set to use the default LAN card address. The default address of the G1369x LAN card is 192.168.254.11 with switches 5&6 up. The switches are located on the card. Set your computer to 192.168.254.10.