Cary 5000: 9:128Write timeout

Hi everybody,

 

after an "out of memory" error we occasionally encounter the "Error code :9:128Write timeout" with multiple instances at the same time. Afterwards, the instrument trys to reset itself, but gets stuck in a loop, unless the measurement software on the PC is closed. While resetting, grinding noises sometimes occur. Does someone know how to solve this issue?

 

Cheers 

Parents
  • Dear Mike,

     

    thank you for your assistance. After reinstalling the cables and powering the system, the initialization of the instrument fails with all kinds of errors:

    9:9614NMP:E9614 8031 timeout

    9:9604NMP:E9604 Microm datum

    9:9328SEQ:E9328 fatal initialisation

    9:9610NMP:E9610 stepper int busy

    9:64 Read time out

    As you said, this does not seem to be easily fixable and a service technician is requested, I appreciate your help though!

    Kind regards 

Reply
  • Dear Mike,

     

    thank you for your assistance. After reinstalling the cables and powering the system, the initialization of the instrument fails with all kinds of errors:

    9:9614NMP:E9614 8031 timeout

    9:9604NMP:E9604 Microm datum

    9:9328SEQ:E9328 fatal initialisation

    9:9610NMP:E9610 stepper int busy

    9:64 Read time out

    As you said, this does not seem to be easily fixable and a service technician is requested, I appreciate your help though!

    Kind regards 