Hi there,
What are possible reasons for these red status? Do you have any idea about that?
Thanks
Hi there,
What are possible reasons for these red status? Do you have any idea about that?
Thanks
Hi
Check above range of values and comments with possible cause...
Hi ,
i decided to change the source according to the table but i am confused.
there are two different codes in manual. Do you have any idea about what are the differences between
source with purge fitting (G8043-67400) and
source assembly (G8043-67505)?
Thanks
Are you using nitrogen purge for system.
Generally, source life is 3 years.
It shows with connect in above pics..
It shows with connect in above pics..
https://community.agilent.com/technical/molecular-spec/f/forum/2750/source-assy-for-ftir-cary-630
Check this discussion for more clarity of purge and without purge...
Hope,Now it is clear to you...
Thank you so much for pictures.I've read this text and I'm signing up for the following question
"Thank you so much for this! The purge is new...I do not see that step in the users guide after replacing the source - it just to reinstall and to check the energy values and adjust the gain if needed. Is there additional instructions for purging I need to read up on? "
That's enough.
Hope, you will get result as per your expectation..
And about reading....
Our whole life will be finish but reading will not be completed...
THis analytical field is limitless...