My cary 5000 UV-VIS-NIR shows Error code 9:9339SEQ NIR wavelength calibration failed and 9:9303SEQ Unable to initialize the instrument.
What is wrong with this machine and how could I fix the problem.
Thanks a lot for your advice in advance.
MK
My cary 5000 UV-VIS-NIR shows Error code 9:9339SEQ NIR wavelength calibration failed and 9:9303SEQ Unable to initialize the instrument.
What is wrong with this machine and how could I fix the problem.
Thanks a lot for your advice in advance.
MK
Hello
From my experience,
1. if the customer is using accessory(like Internal/External DRA, UMA etc.., you should remove it first.
2. Check the shutter of both sides in sample compartment. (you can check it by setting 7-digit dial on main board)
3. set the dial as 15 then check the zero-order light. If the timing belt is not installed correctly, 2 grating will make different diffraction angle. Align it, if it is not good.
B.Regards
CSH
Hello
From my experience,
1. if the customer is using accessory(like Internal/External DRA, UMA etc.., you should remove it first.
2. Check the shutter of both sides in sample compartment. (you can check it by setting 7-digit dial on main board)
3. set the dial as 15 then check the zero-order light. If the timing belt is not installed correctly, 2 grating will make different diffraction angle. Align it, if it is not good.
B.Regards
CSH