Mass Hunter Qualitative analysis by custom Library issue

Hello all,

I am new to the mass hunter qualitative analysis software. I am trying to identify some compounds by creating my own custom library. I have created the library in PDCL by giving the mass of compound, formula of compound and compound name. But when I tried to do the analysis with it, the software is not showing any matched compound names. It does processing for some time but then in the compounds window, nothing is showing. It is not saying that nothing is matched. Even I am sure that the compounds which are in the library should match the samples but still its not showing anything. What could be the possible errors?

Parents Reply Children
No Data
Was this helpful?