Mass Hunter Qualitative analysis by custom Library issue

Hello all,

I am new to the mass hunter qualitative analysis software. I am trying to identify some compounds by creating my own custom library. I have created the library in PDCL by giving the mass of compound, formula of compound and compound name. But when I tried to do the analysis with it, the software is not showing any matched compound names. It does processing for some time but then in the compounds window, nothing is showing. It is not saying that nothing is matched. Even I am sure that the compounds which are in the library should match the samples but still its not showing anything. What could be the possible errors?

Parents
  • Hello  ,

    To troubleshoot the issue you will need to first verify you are able to find the compound and then troubleshoot the identification. Find by Formula requires a peak be integrated for the expected accurate mass EIC. Find by Molecular Feature requires that the 3D signal match your method criteria. I would start by manually making certain there is signal for your expected molecular ion and going from there.

    The Qual Familiarization Guide has some good general usage instructions and also some examples of using Find by Formula and Find by Molecular Feature. It should be available on your MassHunter PC as part of the LCMS Data Analysis Resources.

Reply
  • Hello  ,

    To troubleshoot the issue you will need to first verify you are able to find the compound and then troubleshoot the identification. Find by Formula requires a peak be integrated for the expected accurate mass EIC. Find by Molecular Feature requires that the 3D signal match your method criteria. I would start by manually making certain there is signal for your expected molecular ion and going from there.

    The Qual Familiarization Guide has some good general usage instructions and also some examples of using Find by Formula and Find by Molecular Feature. It should be available on your MassHunter PC as part of the LCMS Data Analysis Resources.