I’m running GRO/DRO/ORO on an 8890 FID. The baseline is cutting across the peaks instead of integrating under the baseline. How do I drop the red integration line so that I get proper area under the given ranges?
I’m running GRO/DRO/ORO on an 8890 FID. The baseline is cutting across the peaks instead of integrating under the baseline. How do I drop the red integration line so that I get proper area under the given ranges?
If this is F.01.03 you can use the word ZERO in the A2 field in Edit Compounds.
If this is F.01.03 you can use the word ZERO in the A2 field in Edit Compounds.
Are using compound type H for the TPH Range?