I'd like to include a value for percent deviation of my ISTD for each of my samples. Can I do this in MS Quant to include in my final report?
Find maintenance, best practice, and troubleshooting articles.
I'd like to include a value for percent deviation of my ISTD for each of my samples. Can I do this in MS Quant to include in my final report?
Hi
for sure you can
first go yo your quant method ---> edit method ---> outlier setup ----> ISTD response and percent deviation
set you percent deviation minimum and maximum
also set your minimum and maximum allowed response if you want
apply them to method and analyze
<<<<<
then add this values to your batch table to review them ( right click on batch table ---> add /remove columns ----> istd compound results ) select what you want to add ( percent deviation , also istd response ratio is recommended )
.... after reviewing you can print your report, you only need to add your istd % dev to your report template
open report builder software and open your template ( for example Gen sample ISTD )
left click on the box which contains all the cells of interest ( it will become dark grey ) then right click and select column
now select istd peak and select istd response percent deviation
>>>>>>>>>
then SAVE AS your template , don't overwrite it
good luck
Hi
for sure you can
first go yo your quant method ---> edit method ---> outlier setup ----> ISTD response and percent deviation
set you percent deviation minimum and maximum
also set your minimum and maximum allowed response if you want
apply them to method and analyze
<<<<<
then add this values to your batch table to review them ( right click on batch table ---> add /remove columns ----> istd compound results ) select what you want to add ( percent deviation , also istd response ratio is recommended )
.... after reviewing you can print your report, you only need to add your istd % dev to your report template
open report builder software and open your template ( for example Gen sample ISTD )
left click on the box which contains all the cells of interest ( it will become dark grey ) then right click and select column
now select istd peak and select istd response percent deviation
>>>>>>>>>
then SAVE AS your template , don't overwrite it
good luck