For some reason a group of us can not open the Mass Hunter Quantitative analysis software. We keep an error code that says "Failed to open batch file. Failed to read binary batch file. We can't analyze any of our bin files.
For some reason a group of us can not open the Mass Hunter Quantitative analysis software. We keep an error code that says "Failed to open batch file. Failed to read binary batch file. We can't analyze any of our bin files.
hello kharrison77,
* does the batch file have a READ only bit?
some team members could do this so that files are not changed.
I think with Masshunter this is not a good idea!
* question:
- so there are several people who have problems, but also people who can successfully open the IDENTICAL batch?
- is this file on a server?
- could it be file access rights that are inherited on the file?
=> batch creator is local admin and the other quant users might be nomal domain users?
=> use file explorer on the batch file => right click > properties > security e.g.
* other idea: use the current QUANT LOG file in "D:\MassHunter\log\Quant".
search inside for lines: "09/11/2020 time [ERROR] "
regards, M.
hello kharrison77,
* does the batch file have a READ only bit?
some team members could do this so that files are not changed.
I think with Masshunter this is not a good idea!
* question:
- so there are several people who have problems, but also people who can successfully open the IDENTICAL batch?
- is this file on a server?
- could it be file access rights that are inherited on the file?
=> batch creator is local admin and the other quant users might be nomal domain users?
=> use file explorer on the batch file => right click > properties > security e.g.
* other idea: use the current QUANT LOG file in "D:\MassHunter\log\Quant".
search inside for lines: "09/11/2020 time [ERROR] "
regards, M.