Is there a way to subtract a TIC or SIM "blank run" chromatogram from your sample's respective TIC or SIM chromatogram in MSD ChemStation? This functionality is in MassHunter and is pretty straightforward. I'm using F.01.00.1903 Enhanced version.
Is there a way to subtract a TIC or SIM "blank run" chromatogram from your sample's respective TIC or SIM chromatogram in MSD ChemStation? This functionality is in MassHunter and is pretty straightforward. I'm using F.01.00.1903 Enhanced version.
It is only possible to subtract a background spectrum from a TIC in MSD ChemStation (File > Subtract Background (BSB) ), not an entire blank run's chromatogram. If this is data being collected by a 2D GC detector (FID, ECD, etc...) then you can use the column compensation feature on the GC to do this, but this feature is not available with MS data.
It is only possible to subtract a background spectrum from a TIC in MSD ChemStation (File > Subtract Background (BSB) ), not an entire blank run's chromatogram. If this is data being collected by a 2D GC detector (FID, ECD, etc...) then you can use the column compensation feature on the GC to do this, but this feature is not available with MS data.
I just wanted to follow-up to see if the previous reply answered your question to your full satisfaction. If so, it would be appreciated if you returned to the thread and marked a "Correct Answer" to increase visibility. If you still require any further help, please let us know and we would be happy to continue assisting you.
I went ahead and flagged the "Correct Answer" on this post. If you still need help though, please let us know and we would be happy to continue working with you.