Hi
Wondering what's wrong with my qualifier parameters.. result table shows no data
thanks
Roy
Find maintenance, best practice, and troubleshooting articles.
Hi
Wondering what's wrong with my qualifier parameters.. result table shows no data
thanks
Roy
Hi Roy,
I updated the tags for better visibility. .
Regards
James
thanks
thank you but that's not it
naturally I updated them..
I guess I can manually integrate them but I don't see why
Hi Roy,
One thing you can try is to delete intermediate files that dictate MH Quant GUI. This is done by deleting folder specified below.
C:\Users\<NT user name>\AppData\Roaming\Agilent_Technologies,_Inc
If this doesn't work you can send batch along with the batch to me [tomi_hamalainen (at) agilent (dot) com] via wetransfer.com or other similar service. I can then run it on my software to confirm that mine behaves the same.
Hi Roy,
are the integrated? Can you go on Manual integration and manual integrate them?
Hi Roy,
are the integrated? Can you go on Manual integration and manual integrate them?
naturally I updated them..
I guess I can manually integrate them but I don't see why
Hi Roy,
One thing you can try is to delete intermediate files that dictate MH Quant GUI. This is done by deleting folder specified below.
C:\Users\<NT user name>\AppData\Roaming\Agilent_Technologies,_Inc
If this doesn't work you can send batch along with the batch to me [tomi_hamalainen (at) agilent (dot) com] via wetransfer.com or other similar service. I can then run it on my software to confirm that mine behaves the same.
Thank you Tomi.
Did not find the file you talk about..
I'm sending you the batch (with embedded method) directory zipped.
FYI - I sent the same batch to my local distributer and he said he does not see a problem.
Perhaps an update is required?
I'm running B.06.00 / Build 6.0.388.0
thanks again
roy
Hi Roy,
Thanks for the files. I double checked the batch and indeed filter settings were most likely okay. What I found out is that you're missing Outlier value for accuracy in the method.
So you want to input values in specific outliers you want to use inside method editor. Here's graphs to show the effect of my adding Accuracy outlier values of 20 %.
Accuracy cells not highlighted for ones that are outside commonly accepted 80-120 % range.
Located Accuracy outlier in method and input value of 20.
Returned to batch table, run Analyze and now outlier is highlighting Accuracy cells properly.
p.s. There's no way to save manual integrations to method. Though when you perform manual integration it will be kept when you integrate/quantitate/analyze the batch,
Hi Roy,
My apologies for not focusing on your initial message regarding missing qualifier values. I tested this and I don't see this issue on my MH Quant B.08.00. Something that I did not have is your reference library so I used NIST14.L during my testing. Could you test it by removing reference library from the method or change it? Also my suggestion is to upgrade to the latest MH Quant version.
thank you.
Is upgrading the version means paying for the software again?
Is it not possible to downlowd an updated version?
Nothing else to try about the qualifiers data? I am using this version for about three month now and never encountered this issue, only recently.. I don't know "what I pressed to make this happen.."
Do you think you will be able to find out more about it if I let you take over the computer via teamviewer (if it's possible for you, of course)?
Funny you mentioned the reference library - I am "stuck" on the method editor with the message error "reference library is required for deconvolution scan"
I do the following to try and resolve it - method -> setup reference library -> I choose a library from my local disc (W9N08.L) and pick a destination.
The program creates the file alright but still no validation..