Struggling with report builder for MassHunter Quant 10

*Edit - report template now attached.

 

I'm trying to configure a quant report for each sample in report builder for MassHunter Quant 10.

 

For each compound I want to display the quant MRM on the LHS, then the two qual MRMs then the ISTD on the RHS. 

 

I've managed to figure out the data binding to get the qualifiers to show up but I am getting the report shown in the picture where each compound shows up twice, with the same wualifier replicated twice on each line.

 

I've looked through the data bindings for the table the graphics sit in and the bindings for the graphics themselves but can't see how I can differentiate between qualifiers.

 

I've spoken to Agilent but apart from the very brief overview there doen't seem to be an instruction manual for report builder.

 

Any help would be greatly appreciated.

 

Cheers

 

Steve

 masshunter quant reportbuilder lc-qqq

attachments.zip
Parents Reply
  • ,

    The error didn't come through, but I know that if any of your compounds have no qualifiers that template will generate an error. It will work for compounds with only one qualifier, but that qualifier will be displayed twice.

     

    If that doesn't sound like what the issue is, please try using the template with some of the shipping data for a test. I know that the QQQ Drugs of Abuse data set will work with this template, but with duplicate qualifiers as I note above. 