MSD G6125B Not Ready

Hello.
I have a problem on several new Agilent 1290 Infinity II instruments without a visible pattern.

 

After a random sample, before starting a new sample. The G6125B mass detector does not transition from the "Not Ready" state to the "Idle" state.
10 minutes pass and the sequence ends due to an error.
The next sequence may also end immediately due to the same error, or it may not complete and work normally.
At the same time, there is no information about the mass detector error in the logs. Only "Instrument nRdy Timeout occured".
A problem may occur a couple of times a day, or it may not occur for several weeks. The problem does not depend on the method used and the sample.
After a short time, everything works.

 

The manual does not say what the "Not Ready" state depends on. According to the displayed data, such as RoughVac and HighVac, everything is in order with the mass detector.

 

The first screenshot captures the moment not long before the completion of the sequence due to an error.

 

Please, help.

 

UPD:
MSD is waiting for temperatures adjusting. But what temperature? The gas and quadrupole temperatures are OK.

attachments.zip
Parents Reply Children
  • The remote cable is connected to the G7112B DAD. Turning the Сhemstation or MSD and other devices on and off does not affect the duration of the operation until the next failure.

     

    UPD: Connecting the remote cable to the multisampler G7167B does not solve the problem. Three of the six new appliances are regularly turned off. Any idea what temperature MSD can expect? So far, it all comes down to data transmission disruption.

     

    UPD2: This is what the problematic devices have in common (only 6 out of 25 have an error): the MSD has firmware version 3.02.50.

Was this helpful?