And if not, are the intensity values of each mass averaged over a time window?
Find maintenance, best practice, and troubleshooting articles.
And if not, are the intensity values of each mass averaged over a time window?
as Mitch described, the average volume of the calibration solution is around that 10 drops / min. However, the actual flow rate depends on several factors - Tightness of the calibration bottle and the screw cap. - Permeability of the CDS valve. - Condition of the rotor seal in the MS Selection Valve. - Condition of the peek capillary to the nebulizer. - permeability of the nebulizer. If the Calbrant is under-promoted, the algorithm of the Tune routine becomes the Multiplier voltage increase nnecessarily to the minimum abundance of the target mass. This usually leads to faster wear of the multiplier. The condition of the nebulizer can be checked very well with the G1960-67470 Nebulizer Adjustment Kit.
Tommy
Translated above with google.
Mitch
as Mitch described, the average volume of the calibration solution is around that 10 drops / min. However, the actual flow rate depends on several factors - Tightness of the calibration bottle and the screw cap. - Permeability of the CDS valve. - Condition of the rotor seal in the MS Selection Valve. - Condition of the peek capillary to the nebulizer. - permeability of the nebulizer. If the Calbrant is under-promoted, the algorithm of the Tune routine becomes the Multiplier voltage increase nnecessarily to the minimum abundance of the target mass. This usually leads to faster wear of the multiplier. The condition of the nebulizer can be checked very well with the G1960-67470 Nebulizer Adjustment Kit.
Tommy
Translated above with google.
Mitch