For the 6400 Series Triple Quadrupole (specifically 6495B), does the checktune inject a fixed volume of calibrant solution every checktune?

And if not, are the intensity values of each mass averaged over a time window?

  • Hi Padraigb,

    All 6400 series LCMS have a calibrant delivery system that supplies calibrant to the nebulizer/ion source at a steady rate. Usually about 7-10 drops/min.

     

    Mitch

  • I believe the flow rate is around ~50ul/min as long as the N2 is adjust properly (13psi) and we are using the Agilent supplied tubing from the divert valve to the nebulizer.

  • Hallo  padraigb

     

    wie Mitch bereits beschrieben hat, ist das durchschnittliche Volumen der Kalibrierlösung um die

    10 Tropfen /min. Jedoch ist die tatsächliche Durchflußmenge von mehren Faktoren abhängig

    - Dichtigkeit der Kalibrierflasche und des Schraubverschlußes.

    - Durchlässigkeit des CDS Ventils

    - Zustand des Rotor Seals im MS-Selection Valve

    - Zustand der Peek Kapillare zum Nebulizer

    - Durchlässigkeit des Nebulizers

     

    Bei zu geringer Förderung des Calbrants wird der Algorithmus der Tune Routine die

    Multiplierspannung unnötigerweise erhöhen, um auf die Mindestabundance der Zielmasse

    zu kommen.

    Dies führt in der Regel zu einem schnelleren Verschleiß des Multipliers.

    Der Zustand des Nebulizers läßt sich sehr gut mit dem G1960-67470 Nebulizer Adjustment Kit.

    beurteilen.

     

    Thommy

     

  • as Mitch described, the average volume of the calibration solution is around that 10 drops / min. However, the actual flow rate depends on several factors - Tightness of the calibration bottle and the screw cap. - Permeability of the CDS valve. - Condition of the rotor seal in the MS Selection Valve. - Condition of the peek capillary to the nebulizer. - permeability of the nebulizer. If the Calbrant is under-promoted, the algorithm of the Tune routine becomes the Multiplier voltage increase nnecessarily to the minimum abundance of the target mass. This usually leads to faster wear of the multiplier. The condition of the nebulizer can be checked very well with the G1960-67470 Nebulizer Adjustment Kit.

     

    Tommy

    Translated above with google.

     

    Mitch

Was this helpful?