Triethylamine Contamination

There is a strong interference ion signal, M/Z=102, and it is suspected to be the pollution of triethylamine. How can we clean it best and effectively? 

Parents Reply
  • You can also try the following:


    1. Rinse the source with isopropanol, water, 25% nitric acid (15'), water (1,5 h),(remove from MS)

    2. Clean the whole ion path , depends on MS 


    You will also need to clean the whole LC.

    3. Exchanging all peek-capillaries
    4. ultrasonic of the rotor seals in isopropanol
    5. Rinse the whole hplc with 25% nitric acid for 15 Minutes
    6. Rinse the whole system with water, purge the pulse damper several times
    7. Rinse the system one night with 50% isopropanol/water with 0.1% FA