Multiple Worklists saving to multiple locations

Hi, 

 

We recently started using column switching on our 6470 QQQ. Essentially we have two differnent methods, phases and columns on the same instrument. We injected two batches sequentially, batch 1 and 2, adding both worklists to the same que, but the data from batch 2 ends up in the same location as batch 1 despite building both worklists offline, and giving each a distinct location.

 

Is there a way to get  the data from batch 1 and 2 to save in seperate locations when they are injected in from the same que?

 

Thanks in advance.

 

John.

Top Replies

Parents
  • John-

     

    Probably the easiest way to accomplish this is to use Study Manager and submit a worklist only study

     

     

    Another option would be to include the different path in the data file name for each sample name in the worklist table.  i.e. specify the desired path in the worklist table

Reply
  • John-

     

    Probably the easiest way to accomplish this is to use Study Manager and submit a worklist only study

     

     

    Another option would be to include the different path in the data file name for each sample name in the worklist table.  i.e. specify the desired path in the worklist table