Hi, I have a question about the procedure of rebooting the computer.
Do I need to operate smart shutdown to the QTOF every time I reboot the computer?
Appreciate your help!
Hi, I have a question about the procedure of rebooting the computer.
Do I need to operate smart shutdown to the QTOF every time I reboot the computer?
Appreciate your help!
Not entirely sure what you mean by "smart shutdown", but rebooting the PC doesn´t affect the QTOF at all. There is a computer with an operating system in the instrument, the SmartCard, so the QTOF can run on it´s own without being permanently controlled by an external PC. However, shutting down the so-called engines (software proceses) is still something I would do before rebooting.
Thanks for your advice! Sorry I meant as operate SmartCard before rebooting.
I was confused some people operate SmartCard before rebooting, and some people don't.
Appreciate your help!
Thanks for your advice! Sorry I meant as operate SmartCard before rebooting.
I was confused some people operate SmartCard before rebooting, and some people don't.
Appreciate your help!