Leaking from needle seat

Hi, I am currently using a 1260 ALS autosampler (6 port valve) on my HPLC system. However, every time I inject sample, or simply run my method at "mainpass" mode for the injection port, I can see a significant pressure drop and also leaking from the needle seat. I just changed a new needle seat, needle, rotor seal and isolation seal. But it is still leaking.

 

Has anyone seen similar problem before? And also, is there a pressure limit for the injection port? My back pressure is around 100 bar. Is that possible that the back pressure is too high that the injection port can not hold this pressure?

 

Thank you!

Parents
  • Hi Chris,

     

    Thanks for your advice! I fixed it based on your instruction.

     

    Do you have an idea how much pressure can this needle hold? I am using the system at around 100 bar back pressure (@5% MeOH). I heard from someone that the needle in the injection port can only tolerate less than 90 bar, is that true? Perhaps I pushed the pressure limit of the needle so that it deformed somehow.

     

    Yanding

Reply
  • Hi Chris,

     

    Thanks for your advice! I fixed it based on your instruction.

     

    Do you have an idea how much pressure can this needle hold? I am using the system at around 100 bar back pressure (@5% MeOH). I heard from someone that the needle in the injection port can only tolerate less than 90 bar, is that true? Perhaps I pushed the pressure limit of the needle so that it deformed somehow.

     

    Yanding