Manuale injection and infinite "acquiring" status with openLab cds

Hello to everybody.

I have an HPLC 1260 infinity II with openLab CDS, with manual injector.

Unfortunately, it happens to me that at the end of the runtime of an injection, the instrument goes to the "idle" status but the injection file stays in the "acquiring" status for an infinite time, without ever going to the "completed" status. Thus the only way I have to do another injection is the "aborted" function but in this way, the file for data analysis is not generated. Please help me to solve this? I am going through the manuals now but I can't figure it out....

Parents
  • Hi, check your method settings. Is the method stoptime set for the pump? If it is set "As injector/no limit" the run goes forever, because the manual injector does not send any signal to the pump that the analysis is over. All other modules can be configured "As Pump/Injector", but the pump has to specify how many minutes the analysis takes.

Reply
  • Hi, check your method settings. Is the method stoptime set for the pump? If it is set "As injector/no limit" the run goes forever, because the manual injector does not send any signal to the pump that the analysis is over. All other modules can be configured "As Pump/Injector", but the pump has to specify how many minutes the analysis takes.