Pump Leak Rate Test Failed

Hello,

I have been conducting Pump leak rate test on 1260 Infinity II (G7112B, 1260 Binary Pump) Module. Lab Advisor version: B.02.10 [253] SP1.

I have tested it several times, but result is always the same (Primary Channel(s) leak(s) somewhere), see Picture 1.

I replaced almost all the spare parts (AIV with cartridges, OBV, Pistons, seals, capillaries, Purge valve and etc.) with exclusionary principle, but problem was still alive. 

Than I changed the configuration to "Without Damper and Mixer", it passes every time, when I do the test, see Picture 2.

What should I check to solve the problem?

Parents Reply Children

  • I have done elasticity capibration (See app.1.).

    Then I did Pump leak rate test four times. Three of them were failure, but one, which was third in a queue has passed with the results near the limit (<3 mkl/min).

    One thing I have considered was that dialog box does not show information about  which solvent should we use during the process.

    I used 2-propanol during the test process. (As I know about old pumps, the solvent is 2-propanol, while this option on Flexible Pump I have is selectable [Water, Organic and etc.])(See App.2.).

    Because of having not any visible evidence on dialog box, I tried to use water instead of 2-propanol twice and the results are excellent (See App.3.).

    To sum up, I got very different results. I have some questions:

    1) I want to know what solvents should I use during the Pump leak rate test or System pressure test.

    2) What type of variables are changed in calculations on Flexible pump when I select water, 2-propanol or some organic and if the calculations are the same between Flexible, Binary 1260 Infinity II and old 1200 pumps.

  • I have fixed this problems on my own.

    As I considered, solvent should have been changed.

    In the MORE... option of The Pump Leak Rate Test you can find that you are able to do test with any solvent, but in Lab Advisor version: B.02.10 [253] SP1  there is no any dialog box to select solvent (I have written this above).

    Finally, I changed Solvent From H2O to IPA and results were satisfied, then I changed it again to H2O and test failed.

    Hence, before you start doing the test, you have to change solvent type from Instrument control menu "Set Solvent/ Compressibility". 

Was this helpful?