How to connect a Sedere Sedex ELSD 100 to an Agilent 1260 HPLC

We have an Agilent Infinity 1260 HPLC system consisting of an autosampler (G7129A) and a pump (G7111A). We would like to connect an ELSD from Sedere (Senex 100LT) to this system. We use OpenLab EZChrom software.
An Agilent technician said that this is possible in principle, but we have problems how to connect the cables for start, autozero and signal with the HPLC. The first thought was to use the Enhanced Remote Interface, but here the connection is not clear/understandable. Does anyone have any ideas or experience with this problem? Many thanks in advance.

Parents
  • I have a Sedex 85 LTII (Shimadzu version) connected to an Agilent 1100 LC system.  We had to purchase an Agilent 1200 Infinity Universal Interface Box II (G1390B).  With the UIB box configured in our OpenLAB CDS LC ChemStation software, I can collect the ELSD signal, trigger A/Z, ELSD Gas on/off, power down, etc.  However, our software cannot control the ELSD parameters; ergo, we must manually input the conditions on the ELSD front panel.

    I can only assume that an UIB box or something similar could interface with OpenLab EZChrom software.

    I get different answers from Agilent on whether there are drivers to control our Shimadzu LT II; but, I gather that is most likely not an option.  May be different for your case.

    Our Technical Sales rep. fleshed out the UIB II configuration and necessary cables for our particular ELSD and Agilent LC.  So, I highly suggest contacting Agilent Tech Support or your Sales person.

Reply
  • I have a Sedex 85 LTII (Shimadzu version) connected to an Agilent 1100 LC system.  We had to purchase an Agilent 1200 Infinity Universal Interface Box II (G1390B).  With the UIB box configured in our OpenLAB CDS LC ChemStation software, I can collect the ELSD signal, trigger A/Z, ELSD Gas on/off, power down, etc.  However, our software cannot control the ELSD parameters; ergo, we must manually input the conditions on the ELSD front panel.

    I can only assume that an UIB box or something similar could interface with OpenLab EZChrom software.

    I get different answers from Agilent on whether there are drivers to control our Shimadzu LT II; but, I gather that is most likely not an option.  May be different for your case.

    Our Technical Sales rep. fleshed out the UIB II configuration and necessary cables for our particular ELSD and Agilent LC.  So, I highly suggest contacting Agilent Tech Support or your Sales person.