RID stabilization

I am having trouble in stabilizing my RID detector. Please provide me proper procedure for stabilization and getting good baseline in RID. Also when i am joining the column or removing the column there is no effect on the RID baseline.

  • I have been having trouble with RID noise and cleaning the system, a new column, insulating the HPLC lines, bypassing the MCGV, elevating the waste & recycle lines, and replacing the in and outlet valves has not reduced it in any real capacity.

    Is the noise caused by G1379A constant or only when the unit is running? The noise on my system is constant (see attached, diluent and standard injection). Can I test whether the noise is caused by the degasser by simply turning it off, or do the lines need to be removed from the unit?

    Thanks for any help

  • Operating an RID without a degasser connected is not a good idea. You could combine two channels of the degasser to see whether the baseline improves, but like I said before, G1379A/B is not a recommended model.

  • Most important - have a proper setup for the RID, otherwise issues e.g. with the baseline, can be expected:

    - never ever operate an RID without a proper degasser. All Agilent degassers can be used with the exception of G1379A and G1379B

    - always have a TCC/MCT in your stack

    - always have to heater in the optical unit turned on. We recommend to set a temperature of 10 degrees above ambient. The higher, the better, but check for the maximum temperature your column can stand. 55 degrees Celsius is the maximum for the RID

    - optical unit temperature and temperature set on the TCC/MCT should be identical

    - use of an isocratic pump is highly recommended. If a quaternary pump is used, bypass the MCGV, if a binary pump is used, pump A and B 50:50 from ONE bottle

    - have the waste and recycle bottles placed above the RID

Was this helpful?