I am having trouble in stabilizing my RID detector. Please provide me proper procedure for stabilization and getting good baseline in RID. Also when i am joining the column or removing the column there is no effect on the RID baseline.
Find maintenance, best practice, and troubleshooting articles.
I am having trouble in stabilizing my RID detector. Please provide me proper procedure for stabilization and getting good baseline in RID. Also when i am joining the column or removing the column there is no effect on the RID baseline.
And as a reminder:
- never ever operate an RID without a proper degasser. All Agilent degassers can be used with the exception of G1379A and G1379B.
- always have a TCC/MCT in your stack
- always have to heater in the optical unit turned on. We recommend to set a temperature of 10 degrees above ambient. The higher, the better, but check for the maximum temperature your column can stand
- Optical unit temperature and temperature set on the TCC/MCT should be identical
I'm not a big fan of building a hood over the detector. In some situations it might be of help, but often the problem lies with either the environmental conditions or an incorrect setup.
BTW: You can trust my answers, I'm the responsible Product Support Engineer for VWD, MWD, DAD, FLD and RID. What I tell here, is exactly what I teach in my trainings.
Why is a G1379A Agilent degasser not compatible with an RID? The RID I'm using is a G7162A.
G1379A and G1379B will work with any Agilent RID (G1362A, G7162A, G7162B), but we don't recommend using them with RID, as their evacuation mode differs from the other degassers and that can be seen on the baseline, which gets more noisy. Depending on your method/peaks, this might be acceptable or not (if you just look for huge peaks, you will very likely not notice the increase in noise).
I have been having trouble with RID noise and cleaning the system, a new column, insulating the HPLC lines, bypassing the MCGV, elevating the waste & recycle lines, and replacing the in and outlet valves has not reduced it in any real capacity.
Is the noise caused by G1379A constant or only when the unit is running? The noise on my system is constant (see attached, diluent and standard injection). Can I test whether the noise is caused by the degasser by simply turning it off, or do the lines need to be removed from the unit?
Thanks for any help
Operating an RID without a degasser connected is not a good idea. You could combine two channels of the degasser to see whether the baseline improves, but like I said before, G1379A/B is not a recommended model.
Most important - have a proper setup for the RID, otherwise issues e.g. with the baseline, can be expected:
- never ever operate an RID without a proper degasser. All Agilent degassers can be used with the exception of G1379A and G1379B
- always have a TCC/MCT in your stack
- always have to heater in the optical unit turned on. We recommend to set a temperature of 10 degrees above ambient. The higher, the better, but check for the maximum temperature your column can stand. 55 degrees Celsius is the maximum for the RID
- optical unit temperature and temperature set on the TCC/MCT should be identical
- use of an isocratic pump is highly recommended. If a quaternary pump is used, bypass the MCGV, if a binary pump is used, pump A and B 50:50 from ONE bottle
- have the waste and recycle bottles placed above the RID
Most important - have a proper setup for the RID, otherwise issues e.g. with the baseline, can be expected:
- never ever operate an RID without a proper degasser. All Agilent degassers can be used with the exception of G1379A and G1379B
- always have a TCC/MCT in your stack
- always have to heater in the optical unit turned on. We recommend to set a temperature of 10 degrees above ambient. The higher, the better, but check for the maximum temperature your column can stand. 55 degrees Celsius is the maximum for the RID
- optical unit temperature and temperature set on the TCC/MCT should be identical
- use of an isocratic pump is highly recommended. If a quaternary pump is used, bypass the MCGV, if a binary pump is used, pump A and B 50:50 from ONE bottle
- have the waste and recycle bottles placed above the RID