The User Manual is not showing in which conditions the Intensity Test is conducted. Do I need to replace the column? What would be the part number?
Find maintenance, best practice, and troubleshooting articles.
The User Manual is not showing in which conditions the Intensity Test is conducted. Do I need to replace the column? What would be the part number?
I guess you mean FLD, not RID?
Ideally, the FLD intensity test is done using the conditions used for wavelength accuracy test and S/N test, so water at 0.25 ml/min. A restriction capillary (2m x0.12mm, PN 5022-2159) should be used.
However, for a quick check it can also be run without flow (but I would make sure that the cell has been flushed with water or IPA).
There are no limits for the intensity test. It's advisable to check for the maximum (for a brand-new FLD this usually ranges from ~16000 to 20000) and a steep increase in the low UV range.